[Halld-pid] ADC
Alexander Ostrovidov
ostrov at hadron.physics.fsu.edu
Fri Sep 30 16:48:13 EDT 2011
On Friday, September 30, 2011, Beni Zihlmann wrote:
> Hi Sasha,
> in the link
> https://halldweb1.jlab.org/talks/2011-3Q/ADC%20spectra%20and%20Time-walk%20corrections_files/
> you show an ADC spectrum of a paddle PMT: adc_l_par_nottrig.gif.
> what is the HV of the PMT set at and what type of ADC are you using?
>
> cheers,
> Beni
> _______________________________________________
> Halld-pid mailing list
> Halld-pid at jlab.org
> https://mailman.jlab.org/mailman/listinfo/halld-pid
>
>
Beni,
This particular plot is done with:
1) CAEN ADC v265 in 12-bit mode. This corresponds to
0.25 pC per count
2) 10-stage PMT with a nominal gain of 1.5*10^6 (at 1750 V).
However, it was used at 2000 V in this particular measurement.
So, its actual gain is unknown in this case.
3) What makes things more complications is a signal attenuation
on its way from PMT anode to the ADC due to a 50-50 splitter,
a 50 Ohm terminator right at the PMT, and 150ft of cable. Right now,
I'm trying to measure what is the correction coefficient I need to apply
to compensate for this attenuation. I should have an estimate of the
number of photoelectrons later today.
Sasha
More information about the Halld-pid
mailing list