[Halld-pid] Minutes, Time-of-Flight Meeting, March 1, 2012
Mark M. Ito
marki at jlab.org
Fri Mar 2 15:53:02 EST 2012
Find the minutes at
https://halldweb1.jlab.org/wiki/index.php/GlueX_TOF_Meeting,_March_1,_2012#Minutes
and below.
_____
GlueX TOF Meeting, March 1, 2012
Minutes
Present:
* FSU: Paul Eugenio, Kurt Koetz, Sasha Ostrovidov, Ian Winger
* JLab: Chuck Hutton, Mark Ito (chair), Tim Whitlatch
Mechanical Design
Note: Chuck brought his PC into the meeting room so we were able to
look at the 3-D I-DEAS drawings.
* The TOF detector modules are clamped between two U-channels that
run along the entire edge of the array.
* Tim proposed lengthening the long scintillators by 1 cm, so that
the edges protrude another 0.5 cm. This is to make the clamp share
load equally between scintillator and light guide.
+ [Added in press] Elton Smith pointed out that it is not a good
idea to put any pressure on the scintillator surface. That
practice promotes crazing. So it looks like in fact we want to
make the counters shorter.
* Tim proposed building a 1/4-scale model of the entire frame,
including some of the detector modules, to see if there are any
assembly issues with the current design.
* Paul proposed building a full-scale model of one corner of the
frame, to test whether mechanical tolerances in the assembled array
are sufficient.
* Kurt expressed some skepticism that the clamping U-channels could
be made with enough precision to both hold the scintillator and
provide a suitable mounting surface for the outer steel magnetic
shields. In fact the current design does not have a provision for
adjusting the angle of the long axis of the shields so that they
miss the PMT's and their bases. That would have to be added.
* Kurt advocated a different approach, where the support for each end
of each module is a separate mechanical assembly, each assembly
supported on a common backbone. These assemblies would allow
adjustment of more of the degrees of freedom and would have greater
precision in doing so. Ian had drawn up a [20]concept some months
ago.
We decided that a lot of issues had been raised that require some study
and further discussion. We agreed to meet again, in a week, Thursday,
March 8, at 10 am (in ARC 428, so Chuck can bring his PC again).
Prototype Status
There was not much to report since the collaboration meeting. The next
step is to measure next measure light transmission through the light
guides.
TDC Choice
Paul repeated his preference for using the CAEN TDC's.
Contract Status
Paul is working on documents to respond the the RFP. Elton sent him
examples from the Indiana contract; Paul has found them helpful.
Retrieved from
"https://halldweb1.jlab.org/wiki/index.php/GlueX_TOF_Meeting,_March_1,_2012"
References
20.
https://halldweb1.jlab.org/wiki/images/8/83/LightGuideMountConcept.pdf
--
Mark M. Ito
Jefferson Lab (www.jlab.org)
(757)269-5295
More information about the Halld-pid