[Halld-pid] Time-of-Flight Meeting Minutes, November 7, 2013
Mark Ito
marki at jlab.org
Fri Nov 8 15:36:33 EST 2013
Folks,
Find the minutes below and at
https://halldweb1.jlab.org/wiki/index.php/GlueX_TOF_Meeting,_November_07,_2013#Minutes
.
-- Mark
________________________________________________________________________________
GlueX TOF Meeting, November 07, 2013
Minutes
Present:
* FSU: Sasha Ostrovidov
* JLab: Hussein Al Ghoul, Eugene Chudakov, Paul Eugenio, Mark Ito
(chair), Aristeidis Tsaris, Beni Zihlmann
* Northwestern: Sean Dobbs
Find a recording of the meeting [32]here.
TOF Delivery
Find some photos [34]here.
* Paul, Hussein and Aristeidis drove 6 crates of modules up from FSU
and stored them in the ESB yesterday. They inspected modules in two
of the crates and they look like they made the trip totally intact.
Each crate holds 8 modules. These are all standard modules. There 5
more crates to transport. Those will come in a couple of weeks.
* FSU also delivered the the steel shields that were made by the
machine shop there. This is all of them.
* When the last 5 crates of modules are transported, FSU will also
bring the collars, stand-offs, and Swiss cheese blocks. All of the
parts are made except for the stand-offs, which are near complete.
* They will also bring equipment to test the modules here, including
a crate, HV supply, and a scope.
* FSU would like to have a small locker or set of shelves in the Hall
or Counting House to store tools, such as the tube removal device.
* We will check with Tim to make sure that floor space in the Hall
for assembly is in his plan. [Added in press: it is.]
* Congratulations to FSU for achieving this important milestone!
Module Testing
Sasha reported that the testing of all 92 modules is now complete,
including the half-length and half-width modules, of course. See
[36]his web page for details. He covered the following topics:
* Status of TOF module tests
* HV balancing for short modules
* Issue with time walk correction fit
Calibration
Sasha led us through the talk he gave last year on calibration,
outlining the thinking at that time. See [38]his slides for details.
The plan is in six steps
* Step 1: TDC / ADC calibration
* Step 2: PMT gain balancing and monitoring
* Step 3: Time-walk corrections
* Step 4: Single-paddle left/right adjustment
* Step 5: Paddle-to-paddle timing offsets
* Step 6: Fit of TOF time to RF time
Sean is pushing each detector group to list the sets of calibration
constants that will be needed. We can then work backwards to the
procedures necessary to measure these constants. Sasha volunteered to
draft a list for us.
Retrieved from
"https://halldweb1.jlab.org/wiki/index.php/GlueX_TOF_Meeting,_November_07,_2013"
References
32. http://cc.readytalk.com/play?id=7rm92
34.
https://plus.google.com/photos/114699847528520444979/albums/5943986330056690545
36.
https://halldweb1.jlab.org/talks/2013-4Q/Status%20of%20TOF%20module%20tests%20%28Nov%207,%202013%29.html
38. https://halldweb1.jlab.org/talks/2013-4Q/tof-2012-10-calib.pdf
--
Mark M. Ito, Jefferson Lab, (757)269-5295, marki at jlab.org
More information about the Halld-pid
mailing list