[Halld-pid] Start Counter issues
Beni Zihlmann
zihlmann at jlab.org
Wed Nov 23 10:24:21 EST 2022
Hi All,
attached you find a note I cobbled together looking at the start counter
data
and made me worry that we are looking at the issue wrong. This may be not
a problem with the path correction but rather a "gain" problem.
I plot the TDC_time minus the RF time VS the signal amplitude. And I do this
for two runs: 73262 (for reference) and a current run 111145.
These plots are generated at three different positions along the z axis at
60cm (middle of the strait section) at 80 cm (bend section) and 92cm
close to the nose (last two plots).
you can (starting from the last plots) that the issue may be the "gain".
in run 73262 the minimum ionizing peak is clearly identified and separated
from the pedestal. This is not so clear for run 111145. Which leads me
to the conclusion that the earlier plots are indicating we cut into the
minimum ionizing peak because the gain is too low.
I will go and look at other runs that are closer to the current run period
with the start counter on. I will also look at what the operating
voltages are
and if they are truly the same as in the past.
Happy Thanksgiving,
Beni
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/halld-pid/attachments/20221123/850160c6/attachment-0001.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: StartCounter_SignalAmplitudesAndTiming.pdf
Type: application/pdf
Size: 100104 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/halld-pid/attachments/20221123/850160c6/attachment-0001.pdf>
More information about the Halld-pid
mailing list