    <font face="Helvetica, Arial, sans-serif">Hi All,<br>
      <br>
      to my dismay I have to report that the issue with the start
      counter timing is most likely<br>
      related to a gain problem with the detector.<br>
      the two attached plots files show the DeltaT = t_paddle - RF_t of
      paddle 2 at two different<br>
      z positions along the beam z=60 (straight section) and z=90 (nose
      section)<br>
      and below the 2d plots is the projection to the horizontal axis.<br>
      <br>
      one can clearly see that the gain is way lower for run 111145
      while for the other runs<br>
      40k to 90k the gain is much larger.<br>
      <br>
      I do not understand why that is because the HV read back is pretty
      much the same for<br>
      all runs. This leads me to believe that either the HV modules are
      lying or somehow the<br>
      connection of the HV to the detector is not done right.<br>
      <br>
      cheers,<br>
      Beni<br>