[Halld-tagger] FW: goniometer quote and design
Kenneth Livingston
Kenneth.Livingston at glasgow.ac.uk
Thu Mar 22 09:47:21 EDT 2012
Hi All,
Here's what I hope is a final design for the goniometer. Still to get a quote and confirmation that we can get all these stages vacuum compatible.
Any questions or comments ?
Cheers,
Ken
________________________________________
From: Dinh, Phong [Phong.Dinh at newport.com]
Sent: 20 March 2012 23:03
To: Kenneth Livingston
Subject: RE: goniometer quote and design
Hello Ken,
Thanks for the files. I've attached a more developed concept drawing with the requested stages. I'm still working with our factory to make sure we can provide the BGS80 and URS100 in vacuum. I don't think this will be an issue, but waiting on confirmation from suppliers for the vacuum compatible components.
So please stay tune and I will keep you updated.
Thanks,
Phong
-----Original Message-----
From: Ken Livingston [mailto:Kenneth.Livingston at glasgow.ac.uk]
Sent: Thursday, March 15, 2012 10:04 AM
To: Dinh, Phong
Subject: Re: goniometer quote and design
Hi Phong,
Thanks for the update.
I think we'll have to go with the MTM100PE1V6 vertical translation,
since there seems to be no other option. However, I've noticed a couple
of other crucial differences between the quoted specifications and those
outlined in my original. I've noted these below as comments after each
axis as described in the quote. I've also attached a pdf made by our own
CAD expert using step file from the Newport web site showing
approximately the layout we'd like. I'm trying to keep the overall
dimensions to a minimum since this will need to go in a vacuum chamber.
I'll be in Jefferson Lab next week and will meet the engineers for the
project and discuss the specs with them. If you can magage to address
any of these issues before that I'd really appreciate it.
Regards,
Ken
Standard Items:
--------------------------------------------------
/· (1) MTM150PE1V6 -- [aka X] Long-travel steel linear stage; Ball
bearing guideway; Full-step motor; 10:1 Gear Reducer;
Backlash-compensated lead screw drive; Shaft-mounted 2000 pts/rev.
optical rotary encoder with index; 150mm travel; 1μm resolution;
1.5μm guar. uni-directional repeatability; 1.5μm guar. reversal value,
5.0μm guar. on-axis accuracy, 40μrad pitch, 55μrad yaw, 1mm/sec
max speed, 4mm/sec² max acceleration
/
**OK**
---------------------------------------------------
/· (1) MTM100PE1V6 -- [aka Y] Long-travel steel linear stage; Ball
bearing guideway; Full-step motor; 10:1 Gear Reducer;
Backlash-compensated lead screw drive; Shaft-mounted 2000 pts/rev.
optical rotary encoder with index; 100mm travel; 1μm resolution;
1.5μm guar. uni-directional repeatability; 1.5μm guar. reversal value,
5.0μm guar. on-axis accuracy, 40μrad pitch, 35μrad yaw, 1mm/sec
max speed, 4mm/sec² max acceleration
/
**OK**
---------------------------------------------------
/· (1) URS50BPPV6 -- [aka rY] Precision rotation stage, Vacuum version,
Hardened steel with aluminum body, Large diameter ball
bearing movement, Ground worm gear with self-compensating leadscrew,
1:80 worm gear ratio, Step motor drive, 0.0014° resolution
(1/2 full step), 0.004° uni-directional repeatability, 0.006°
bi-directional repeatability, 0.04° absolute accuracy, 50μrad wobble, 6μm
eccentricity, 20°/sec max speed, 80°/sec² max acceleration, travel range
360° continuous, 50N centered load capacity, 3m interconnect
cable, 1.7kg stage weight/
** This needs to have a 50mm diameter hole in the center. This is a
crucial requirement. We'd need a vacuum compatible version of URS100BPP,
or even URS150BPP **
** Is there no encoder on this axis?
---------------------------------------------------
/· (1) URS75BPPV6 -- [aka rZ] Precision rotation stage, Vacuum version,
Hardened steel with aluminum body, Large diameter ball
bearing movement, Ground worm gear with self-compensating leadscrew,
1:90 worm gear ratio, Step motor drive, 0.0002° resolution
(1/100 full step), 0.004° uni-directional repeatability, 0.014°
bi-directional repeatability, 0.03° absolute accuracy, 50μrad wobble, 3μm
eccentricity, 20°/sec max speed, 80°/sec² max acceleration, travel range
360° continuous, 100N centered load capacity, 3m
interconnect cable, 1.7kg stage weight
/
** OK **
---------------------------------------------------
/· (1) BGM120PEV6 -- [aka rX]Goniometric cradle, Stainless steel with
aluminum body, Ball bearing movement, Ground worm gear,
1:180 gear ratio, Full-step motor drive, Shaft-mounted 2000 pts/rev.
optical rotary encoder, 0.001° resolution, 0.004° uni-directional
repeatability, 0.02° reversal value, 0.05° on-axis accuracy, 600μrad
wobble, 2°/sec max speed, 8°/sec² max acceleration, travel range
±45°
/
** We need the 0.0002 resolution on this stage. This is a crucial
requirement. Eg. vacuum compatible version of
http://search.newport.com/?q=*&x2=sku&q2=BGS80PP **
> Hello Ken,
>
> I just wanted to follow up with you. I believe the quote should have
> been generated for you. If you have any questions about the quote,
> please let me know and I can assist.
>
> In regards to the GTS30V, I have spoke with our factory and we won't
> have the capabilities to find the correct components that would allow
> this stage to be vacuum compatible at 10-6. So I believe our option
> would remain the same as what is quoted. I've attached the concept
> drawing again, but this time with the vertical set to the side.
>
> One thing I wanted to find out if you had a time frame that was needed
> for the system? I know the quote states 12 weeks, but this could be
> pulled in.
>
> Best regards,
>
> Phong
>
> **
>
> **
>
> ------------------------------------------------------------------------
>
> *From:*Ken Livingston [mailto:Kenneth.Livingston at glasgow.ac.uk]
> *Sent:* Thursday, March 08, 2012 5:50 AM
> *To:* Dinh, Phong
> *Subject:* Re: goniometer quote and design
>
> Hi Phong,
> Thanks. This is very useful.
> A vacuum level of 10-6 torr is fine.
>
> I would appreciate a quote for the parts with assembly, and an update
> when you get some information on a vacuum version of the GTS30V
> <http://search.newport.com/?q=*&x2=sku&q2=GTS30V>.
>
> We managed to find step files for some, but not all of the components.
> Is there a single download somewhere with a library of the components?
> Or a directory on the web?
>
> Regards,
> Ken
>
>
> On 03/08/2012 01:42 AM, Dinh, Phong wrote:
>
> Hello Ken,
>
> I've listed the stages below that I believe will be potential for this
> system. And they are linked to our website, from there you can get the
> STEP files.
>
> Axis 1: MTM150PE1V6
> <http://search.newport.com/?q=*&x2=sku&q2=MTM150PE1BV6>
>
> Axis 2: MTM150PE1V6
> <http://search.newport.com/?q=*&x2=sku&q2=MTM150PE1BV6>
>
> Axis 3: UTS50PPV6 <http://search.newport.com/?q=*&x2=sku&q2=URS50BPPV6>
>
> Axis 4: UTS75PPV6 <http://search.newport.com/?q=*&x2=sku&q2=URS75BPPV6>
>
> Axis 5: BGM120PEV6 <http://search.newport.com/?q=*&x2=sku&q2=BGM120PEV6>
>
> The XPS-C6 <http://search.newport.com/?q=*&x2=sku&q2=XPS-C6> would be
> the recommended controller and the driver cards would be (5) XPS-DRV01
> <http://search.newport.com/?q=*&x2=sku&q2=XPS-DRV01>.
>
> With the XPS controller the communication is now through Ethernet. The
> XPS does have drivers for other software such as LabVIEW, Matlab.
>
> Also it does contain a GUI and you can develop TCL scripts. Many
> options and you can find more details from the link under Downloads.
>
> I do have a question, as I have been assuming vacuum level at 10-6
> Torr, is this correct?
>
> If you do need a quote for the above parts with assembly, I can
> provide that for now until I can gather more information about a
> vacuum version of the GTS30V
> <http://search.newport.com/?q=*&x2=sku&q2=GTS30V>.
>
> Best regards,
> Phong
>
> -----Original Message-----
> From: Ken Livingston [mailto:Kenneth.Livingston at glasgow.ac.uk]
> Sent: Wednesday, March 07, 2012 1:51 AM
> To: Dinh, Phong
> Subject: Re: goniometer quote and design
>
> Hi Phong,
>
> We discussed the drawing you sent, but like me the others wanted more
>
> details. Ideally, we'd like to get rid of the very large vertical stage
>
> which would dictate the size of the vacuum chamber.
>
> I'd like to know which motor stages from the Newport Range are in your
>
> current design, so that we can look at specifications and prices. Can
>
> you please let me have that information. As I said in the previous mail,
>
> it would be ideal if we had access to the AUTOCAD/STEP files. :
>
> "We have an autocad specialist in the group, who has developed similar
>
> designs in the past using the autocad / step files which are available
>
> for the THORLABS motion stages. Ideally we'd like to do the
>
> configuration ourselves in autocad, using your design as a starting
>
> point. Would it be possible for you to give us the information we
>
> require to do this? eg step files for the different motor stages and a
>
> recommendation of the possible stages we can use, and the details of
>
> those which are in your drawing."
>
> Controller.
>
> Can you let me know which controller you advise for this system.
>
> We want to be able to send sequences of commands from our own computer
>
> programs - currently we do this on an RS-232 line using MM2000 type
>
> controllers.
>
> Regards,
>
> Ken
>
> On 03/07/2012 01:46 AM, Dinh, Phong wrote:
>
> > Hello Ken,
>
> >
>
> > I hope your meeting went well and I'm curious of the feedback you
>
> > received.
>
> >
>
> > For the vertical stage in the back, this would be easily moved to the
>
> > side to avoid the beam.
>
> >
>
> > I'm still working with our factory to see if we can provide a vacuum
>
> > compatible stage such as the VP-25X or the GTS30V. If this is possible,
>
> > then as you mentioned we can would placed this between axis 3 and 4.
>
> >
>
> > Looking forward to your feedback.
>
> >
>
> > Thanks,
>
> >
>
> > Phong Dinh
>
> > Applications Engineer
>
> > Newport Motion
>
> >
>
> > Newport Corporation
>
> > 1791 Deere Ave.
>
> > Irvine, CA 92606
>
> > Office: (949) 253-1819
>
> > www.newport.com <http://www.newport.com>
>
> >
>
> > For Motion, Think Newport!
>
> >
>
> >
>
> > -----Original Message-----
>
> > From: Ken Livingston [mailto:Kenneth.Livingston at glasgow.ac.uk]
>
> > Sent: Friday, March 02, 2012 3:23 AM
>
> > To: Dinh, Phong
>
> > Cc: Anthony Clarkson; Peter Lumsden
>
> > Subject: Re: goniometer quote and design
>
> >
>
> > Hi Phong,
>
> > Thanks a lot for the design drawing. I'll discuss them at monday's
>
> > meeting. That's already pretty close to what we're after.
>
> >
>
> >
>
> > For ease of reference I've put some numbered labels on the axes and
>
> > attached your drawing, together with my original with the same axis
>
> > numbering. I've also put in a sketch of the kind of radiator wheel we'd
>
> > have mounted in the center.
>
> >
>
> > I have a few comments and questions:
>
> >
>
> > 1. I'd like to avoid having that big, heavy, vertical translation
>
> > stage, if possible. See later comment. Also, in your current version
>
> > that would block the beam - but I guess it could easily be repositioned.
>
> >
>
> > 2. There's some flexibility in ordering of the stack. In particular, the
>
> >
>
> > ordering of pitch and yaw is not important. So your change there is
>
> > fine. We should take whatever order is easiest.
>
> >
>
> > 3. Let me clarify what we need to be able to do, since I wasn't very
>
> > specific about distances and angles:
>
> > a. We need to be able select one radiator from the disc mounted within
>
> > the ~50mm centre ring which the beam goes through.
>
> > b. We need to be able to rotate this selected radiator by up to +/-
>
> > 45deg to get it in the azimuthal orientation we require.
>
> > c. We then need to be able to translate in x and y to get this radiator
>
> > into the beam. Translations of about 0-25mm in y and +/-25 in x.
>
> > d. Once a crystal radiator is in the beam line we make adjustments of
>
> > typically 0 - 1 deg on pitch and yaw to get the crystal in the "zero
>
> > position", where that lattice is parallel to the beam. We then make very
>
> >
>
> > small movements (eg 0.05 deg steps) on pitch and yaw to adjust the angle
>
> >
>
> > between the beam and the crystsal lattice.
>
> > e. we also need to be able to translate the whole gonoimeter out of the
>
> > beamline (eg 0-150mm).
>
> >
>
> > This gives some flexibility in the ordering of the axes:
>
> > a. Axis 1, horizontal translation should be at the bottom of the stack
>
> > since it will need to translate the whole goniometer
>
> > b. Axis 3, roll, should be at the top of the stack, since it will rotate
>
> >
>
> > by +-45 - and potentially +-170.
>
> > c. Axes 4 and 5, yaw and pitch, have typically very small motion away
>
> > from home position (eg 1deg), so their ordering is not important. We can
>
> >
>
> > do whatever is easiest.
>
> > d. Axis 2, vertical translation of 0-25mm can really be anywhere between
>
> >
>
> > axes 1 (horizontal translation) and axis 3 (roll).
>
> > From your diagram it looks to me like the easiest for axis 2 would be a
>
> >
>
> > small platform in between axes 4 and 3. For example, something like
>
> > VP-25X, or GTS30V.
>
> >
>
> >
>
> > We have an autocad specialist in the group, who has developed similar
>
> > designs in the past using the autocad / step files which are available
>
> > for the THORLABS motion stages. Ideally we'd like to do the
>
> > configuration ourselves in autocad, using your design as a starting
>
> > point. Would it be possible for you to give us the information we
>
> > require to do this? eg step files for the different motor stages and a
>
> > recommendation of the possible stages we can use, and the details of
>
> > those which are in your drawing.
>
> >
>
> > Cheers,
>
> > Ken
>
> >
>
> >
>
> >
>
> > On 03/01/2012 07:56 PM, Dinh, Phong wrote:
>
> >> Hello Ken,
>
> >>
>
> >> Yes, I have reviewed this and the stack setup will need to be
>
> > different
>
> >> than the stack you provided since we don't have a vertical stage
>
> > similar
>
> >> to the drawing with 25mm and can handle the load of the other stages.
>
> >>
>
> >> I've attached just a concept drawing (not fully completed) to give you
>
> >> an idea. As you will see, for the vertical translation, we will have
>
> > to
>
> >> mount a stage in back of the rotations to give you the min 25mm
>
> > travel.
>
> >> Also, I wanted to check on the stack of rotation axes. We placed the
>
> >> vertical rotation(yaw) on top of the horizontal rotation(Pitch). I
>
> >> wanted to check with you if this will cause any issues with your
>
> >> movement sequences. The reason behind this, is due to the standard
>
> >> vacuum stages and the limited of standard size we have.
>
> >>
>
> >> If this is an issue, I can always request and modify our other stages
>
> > to
>
> >> be vacuum compatible, but the lead time and pricing will reflect.
>
> >>
>
> >> Please let me know your feedback.
>
> >>
>
> >>
>
> >>
>
> >> Best regards,
>
> >> Phong
>
> >>
>
> >>
>
> >> -----Original Message-----
>
> >> From: Ken Livingston [mailto:Kenneth.Livingston at glasgow.ac.uk]
>
> >> Sent: Thursday, March 01, 2012 5:58 AM
>
> >> To: Dinh, Phong
>
> >> Subject: Re: goniometer quote and design
>
> >>
>
> >> Hi Phong,
>
> >> Is there any progress with this goniometer request. I have a meeting
>
> > on
>
> >> Monday where I need to give an update on the status.
>
> >>
>
> >> Regards,
>
> >> Ken
>
> >>
>
> >> On 02/22/2012 12:55 AM, Dinh, Phong wrote:
>
> >>> Hello Ken,
>
> >>>
>
> >>> My name is Phong Dinh and your inquiry about a 6 axes system was
>
> >>> forward to me from Mark White. I will be glad to assist you.
>
> >>>
>
> >>> I received the PDF file and also the Excel file that you provide and
>
> > I
>
> >>> am somewhat familiar with the stages.
>
> >>>
>
> >>> Currently these stages are obsolete so we will have to pick from the
>
> >>> current inventory of stages we have now.
>
> >>>
>
> >>> I do not believe this will be a problem, as we should have comparable
>
> >>> stages. I will look into this and provide you with a proposal and
>
> >> quote.
>
> >>> However, since the PDF only consisted the drawing of the 5 axis
>
> > system
>
> >>> and the files that I have of the older system seem to be corrupted. I
>
> >>> am not able to figure out where the 6 axis needs to be.
>
> >>>
>
> >>> Will it be possible for you to let me know where you would like the
>
> >>> additional translation stage to be? Perhaps a sketch of each axis
>
> > will
>
> >>> be helpful and your coordinate system.
>
> >>>
>
> >>> If you have any other questions, please let me know.
>
> >>>
>
> >>>
>
> >>> Look forward to hearing from you.
>
> >>>
>
> >>> Thanks,
>
> >>>
>
> >>> *Phong Dinh**
>
> >>> **Applications Engineer**
>
> >>> **Newport**Motion***
>
> >>>
>
> >>> *Newport**Corporation**
>
> >>> **1791 Deere Ave.**
>
> >>> **Irvine**, CA 92606**
>
> >>> **Office: (949) 253-1819**
>
> >>> www.newport.com <http://www.newport.com><http://www.newport.com> *
>
> >>>
>
> >>> //**//
>
> >>>
>
> >>> //
>
> >>>
>
> >>> **/Newport Corporation Family of Brands -/**/Corion(r) //*//New
>
> >> Focus(tm)
>
> >>> //*//Oriel(r) Instruments //*////RichardsonGratings(tm)
>
> >> //*//Spectra-Physics(r) /
>
> >>> //
>
> >>>
>
> >>> /This message and any attachments thereto are intended only for the
>
> >>> designated recipient(s) and may contain confidential or proprietary
>
> >>> information and be subject to the attorney-client privilege or other
>
> >>> confidentiality protections. If you are not a designated recipient,
>
> >>> you may not review, use, copy or distribute this message or any
>
> >>> attachments. If you receive this in error, please notify the sender
>
> >>> by reply e-mail and permanently delete the original and any copies of
>
> >>> this message and any attachments thereto. Thank you./
>
> >>>
>
> >
>
--
=======================================================
Ken Livingston
Dept. of Physics& Astronomy, Tel: +44 141 330 6428
University of Glasgow, Fax: +44 141 330 5889
Glasgow G12 8QQ.
Scotland. UK.
=======================================================
-------------- next part --------------
A non-text attachment was scrubbed...
Name: 11SI71243.jpg
Type: image/jpeg
Size: 304551 bytes
Desc: 11SI71243.jpg
Url : https://mailman.jlab.org/pipermail/halld-tagger/attachments/20120322/0bc34d8d/attachment-0001.jpg
More information about the Halld-tagger
mailing list