[Halld-tagger] specs for microscope readout power supplies

Hovanes Egiyan hovanes.egiyan at gmail.com
Mon Dec 9 10:51:18 EST 2013


Hi Richard,
what type of controls is needed for these? Do we need to be able to 
remotely
control them, turn them on/off, power cycle them?
Hovanes.

On 12/09/2013 10:37 AM, Fernando J Barbosa wrote:
> Hi Richard,
>
> OK, looks good. I will look at options for the supplies.
>
> Thanks and best regards,
> Fernando
>
> On 12/9/2013 10:28 AM, Richard Jones wrote:
>> Fernando,
>>
>> Our background study showed that we could continue to function for 10 
>> years of high-rate running without annealing, just allowing the dark 
>> rate to increase at the rate that was measured in Hall B and allowing 
>> the supply current to increase.  Under the zero-shielding scenario, 
>> just leaving the electronics unshielded in the tagger hall, we 
>> estimated that we would reach the point where we would need to do 
>> something (eg. anneal or replace the SiPM's) after 10 years of 
>> high-rate running.  The usual assumptions were applied (100 days of 
>> running per year, standard 20 micron diamond, 12 GeV electrons at 2.2 
>> microAmps).  With shielding, we expect to reduce this figure by a 
>> factor of 5-10 (the simulation gave a factor around 8), so we should 
>> be able to run for the duration of GlueX without annealing.
>>
>> I agree with you that we can segment the supplies.  Each of them can 
>> be split up to 6 ways, because there are 6 identical backplanes in 
>> the system.  The specs we posted were for the total sum of all 6 
>> backplanes.
>>
>> -Richard J.
>>
>>
>> On Mon, Dec 9, 2013 at 9:52 AM, Fernando J Barbosa <barbosa at jlab.org 
>> <mailto:barbosa at jlab.org>> wrote:
>>
>>     Hi Richard,
>>
>>     Certainly the rates in the tagger are higher but what is your
>>     planning for annealing the SiPMs? I am just curious given the
>>     higher rates.
>>
>>     Regarding the power supplies, you mentioned the possibility of
>>     having three supplies instead of one for the whole system. Is it
>>     possible to consider further segmentation? One channel per
>>     backplane? Do you have a document that shows the whole system as
>>     installed? The issue is also related to the space available under
>>     the magnet.
>>
>>     Best regards,
>>     Fernando
>>
>>
>>
>>     On 12/6/2013 10:09 AM, Richard Jones wrote:
>>>     Fernando,
>>>
>>>     These estimates for the BCal do not apply to the TAGM, where the
>>>     rates are much higher.  We can set up a meeting to discuss this
>>>     further if you feel we need it.  We worked out these rates back
>>>     during the design phase for the readout electronics, and had
>>>     them reviewed within the photon beam working group. Also keep in
>>>     mind that the neutron flux is much higher in the tagger hall
>>>     than in the experimental hall.  On the other hand, the TAGM can
>>>     operate efficiently at much higher single-pixel rates than can a
>>>     calorimeter because it is not producing an energy measurement,
>>>     but only time, and the photon statistics are high, <n> = 350-400
>>>     within a 15ns window.  We are shielding these electronics, but
>>>     we only expect to cut down the neutron flux by about a factor
>>>     10-20 in this way.  Alex Somov did the neutron rates and
>>>     shielding studies for us, and can provide more details in this
>>>     regard.
>>>
>>>     -Richard J.
>>>
>>>
>>>     On Fri, Dec 6, 2013 at 9:04 AM, Fernando J Barbosa
>>>     <barbosa at jlab.org <mailto:barbosa at jlab.org>> wrote:
>>>
>>>         Hi Richard,
>>>
>>>         In reference to the BCAL and taking Yi's measurements in
>>>         Hall A regarding radiation damage, we settled on a maximum
>>>         of 10mA for 10 arrays or 160 cells (3mm x 3mm). This works
>>>         out to 62.5uA per cell and your estimate is higher by a
>>>         factor greater than 10. What is your plan for annealing the
>>>         SiPMs? Frequency?
>>>
>>>         Did you look into any supplies that might fit your
>>>         requirements? I recall that each backplane of the TAGM has a
>>>         single bias supply input, correct? This would be important
>>>         in considering a multi-channel supply system. Please send me
>>>         your latest drawings on the TAGM system and its installation
>>>         in the Tagger hall.
>>>
>>>         Best regards,
>>>         Fernando
>>>
>>>
>>>
>>>         On 12/5/2013 10:50 PM, Richard Jones wrote:
>>>>         Hello Fernando,
>>>>
>>>>         I am spec'ing the Vbias supply capacity at 500mA to allow
>>>>         for the maximum SiPM draw current that is permitted by the
>>>>         bias voltage supply system.  Initially we expect the
>>>>         average current to be around 15 mA at full intensity (2.2
>>>>         uA electrons on 20 micron diamond) but that it will
>>>>         increase gradually over time due to radiation damage.  To
>>>>         get the maximum current spec, I imagine the worst possible
>>>>         circumstances under which the radiation damage might
>>>>         accumulate faster than expected, and ask how much current
>>>>         could we sustain across all channels before the resolution
>>>>         and efficiency degrade significantly due to dark current.
>>>>          The answer to that is approximately 800uA per channel,
>>>>         which totals about 400mA from the supply.  I rounded that
>>>>         up to 500mA.
>>>>
>>>>         -Richard Jones
>>>>
>>>>
>>>>         On Thu, Dec 5, 2013 at 4:51 PM, Fernando J Barbosa
>>>>         <barbosa at jlab.org <mailto:barbosa at jlab.org>> wrote:
>>>>
>>>>             Hi Richard,
>>>>
>>>>             I looked at the specs and the bias supply current seems
>>>>             to be excessive at 500mA. What's the reason for this?
>>>>
>>>>             Best regards,
>>>>             Fernando
>>>>
>>>>
>>>>
>>>>             On 12/5/2013 3:33 PM, Richard Jones wrote:
>>>>
>>>>                 Hello Alex,
>>>>
>>>>                 Here is a draft of the specs for the microscope
>>>>                 readout dc power supplies.  They can be a single
>>>>                 unit for each of 3 levels, or a few lower-capacity
>>>>                 supplies in tandem with the same total output
>>>>                 capacity.  Please request justification for any of
>>>>                 the specs.  One thing we do not spec here, but
>>>>                 think would be useful, is separate delivery and
>>>>                 sense terminals for each output.  We have separate
>>>>                 pins on the backplane connector for this purpose.
>>>>
>>>>                 http://zeus.phys.uconn.edu/wiki/index.php/Microscope_Electronics#Power_Supply_Requirements
>>>>
>>>>                 -Richard Jones
>>>>
>>>>
>>>>
>>>
>>>
>>
>>
>
>
>
> _______________________________________________
> Halld-tagger mailing list
> Halld-tagger at jlab.org
> https://mailman.jlab.org/mailman/listinfo/halld-tagger

-------------- next part --------------
An HTML attachment was scrubbed...
URL: https://mailman.jlab.org/pipermail/halld-tagger/attachments/20131209/69edf5a4/attachment.html 


More information about the Halld-tagger mailing list