    Fernando,<br>
    <br>
    Alex has already sent you a proposed naming scheme for the cables
    connecting the tagging counters to the electronics racks.&nbsp; Here is
    some further elaboration to explain how the cabling should be
    configured.&nbsp; It applies to the microscope (MIC-xxx), and perhaps to
    the fixed array (FA-xxx) channels as well.<br>
    <ol>
      <li>There should be only one coax cable per electronic signal
        running from the detector to the electronics racks.&nbsp; This signal
        will be split at the electronics racks, and short cables run
        from the splitter outputs to the discriminator input (split
        output 1) and the fadc input (split output 2).</li>
      <li>These short cables should be long enough to span the height of
        one rack, but not much more.</li>
      <li>The names requested by Alex, MIC-T-nnn and MIC-A-nnn, are only
        appropriate for these short cables that come after the
        splitter.&nbsp; Before the splitter, these two are one and the same
        signal, and should be distinguished by names of their own.<br>
      </li>
      <li>For the long cables from the microscope to the racks (before
        the splitter), I suggest the following naming scheme: MIC-S-nnn
        where nnn runs from 1 to 122.</li>
    </ol>
    <p>I also echo Liping's desire to have a few extra cables run along
      the long route from detector to racks, just because it would be
      disruptive to try to replace a few broken cables once they are
      laid in the trays.&nbsp; I leave to your discretion how many extra
      would be desirable.&nbsp; Perhaps there is a policy for other detectors
      that you could apply to us.&nbsp; For these, one might just extend the
      above scheme, for example letting nnn go from 1 to 135 instead of&nbsp;
      1 - 122.<br>
    </p>
    <p>-Richard Jones<br>
    </p>
    <br>
    <br>
    <br>
    <br>
    <div class="moz-cite-prefix">On 6/21/2012 9:41 AM, Alexander Somov
      wrote:<br>
    </div>
    <blockquote cite="mid:Pine.LNX.4.64.1206210935090.19313@jlabl1.jlab.org" type="cite">
      <pre wrap="">
Hi Fernando,


Here are cabling schemes for the PS, the  microscope,
and the fixed-array counters. I would not put fadc
channel on labels. If some fadc channels are dead, the
cabling is screwed up. I assume that labels are the same
on both cable ends; we don't have many channels in
these detectors.

Cheers,
        Alex


P.S. Richard, Franz, please take a look on the microscope
and fixed array labling




Pair Spectrometer
-----------------

DET - SIDE - FADC/TDC - CH

1. Fine Granularity (128 + 128 = 256 channels)

PS1 - 1 - A - 1
..............
PS1 - 1 - A - 256


PS1 - 2 - A - 1
..............
PS1 - 2 - A - 256

2. Low granularity ( 8 + 8 = 16 channels)

PS2 - 1 - A - 1
...............
PS2 - 1 - A - 8


PS2 - 1 - T - 1
...............
PS2 - 1 - T - 8


PS2 - 2 - A - 1
...............
PS2 - 2 - A - 8


PS2 - 2 - T - 1
...............
PS2 - 2 - T - 8



Fixed Array (190 channels)
--------------------------

DET - FADC/TDC - CH

FA - A - 1
............
FA - A - 190


FA - T  - 1
............
FA - T  - 190


Microscope (120 + 2 = 122 ch)
-----------------------------

DET - FADC/TDC - CH

MIC - A - 1
...........
MIC - A - 122


MIC - T - 1
...........
MIC - T - 122