Hi Fernando
the point of lowering the gain was to increase the dynamic range so
we could accommodate
a larger range of pulse heights (which was crucial for dE/dx). Based on
the noise that we were
seeing, we carefully worked out that given the same noise and maximum
signal size,
that a gain of 3 to 5 lower would allow us to do dE/dx except for the
extremely slow protons.
If the maximum has actually decreased by a factor of two, then we have
effectively lost that
in the dynamic range and I think that we probably have a problem with
this configuration.
Curtis
On 6/10/10 8:24 PM, Fernando J. Barbosa wrote:
> Hi Naomi,
>
> Now it makes sense. The original preamp had a gain of 1.5 mV/fC and
> the new one that you have is 0.57 mV/fC. So, a factor of 2.63 lower
> now. This seems to be confirmed on the first set of plots if you were
> to cut the "overflows" from the left plot (just by looking at the plot).
>
> The dynamic range on the old preamp was wider with 400 fC at the 1%
> level; the new preamp is at 380 fC @ 5% (for your configuration). The
> older preamp had a much wider DR with a smooth roll off before hard
> saturation; the new preamp has narrower DR before hard saturation.
> It is important to keep in mind that we want to operate in the linear
> region, perhaps even better than the 5% level. This is seen on the
> next sets of plots for the new preamp has a hard saturation at ~2100
> channels. The fADC125 will be optimized to have this mapped to the
> upper level of its 12-bit range.
>
> I cannot say much about the energy plots because of non-linearity and
> saturation on both old and new preamps. The detector gain is too high
> and I suggest that you take additional runs starting at 1750V.
>
> Thanks and best regards,
> Fernando
>
>
>
>
> Naomi Jarvis wrote:
>>
>> Hello,
>>
>> I just wanted to correct something I wrote yesterday (which did not
>> make sense anyway). From looking at the scope traces with the 55Fe
>> source I thought that the new preamp gain was less than that of the
>> old preamp. After actually collecting data and extracting amplitude
>> histograms it turned out that the preamp output was in fact
>> saturating, although I could not see this on the scope. I thought I
>> could distinguish between saturated and non-saturated traces, but
>> evidently this was not the case. So I have added some 55Fe plots to
>> the wiki:
>>
>> http://www.jlab.org/Hall-D/software/wiki/index.php/Preamp_4:_Comparison_of_55Fe_data_with_original_preamp
>>
>>
>> These are again comparing the two preamps, and to be clear, the only
>> difference between the two sets of runs is that the preamps are
>> switched and the baseline on the shaper is set to approx -130mV
>> (because the preamps have a different baseline). Everything else is
>> the same, both sets of data were taken on the same day.
>>
>> Is there anything else you would like me to try with it tomorrow, or
>> leave counting over the weekend?
>> Naomi.
>>
>>
>>
>> Fernando J. Barbosa wrote:
>>> Hi all,
>>>
>>> It seems that Monday @ 10 is preferable to most.
>>>
>>> Gerard, I will discuss with you any issues related to the preamps
>>> and ASICs afterwards.
>>>
>>> In preparation for the meeting, I suggest reviewing GlueX docs 1070,
>>> 1079 and 1364. I will send an email with the agenda for the meeting
>>> by the end of the week.
>>>
>>> Thanks and best regards,
>>> Fernando
>>>
>>> Curtis A. Meyer wrote:
>>>> Monday at 10:00am would be great! Curtis
>>>> On 6/9/10 2:31 PM, Fernando J. Barbosa wrote:
>>>>> Hi Curtis,
>>>>>
>>>>> Would Monday @ 10 next week work for everyone?
>>>>>
>>>>> Best regards,
>>>>> Fernando
>>>>>
>>>>>
>>>>>
>>>>> Curtis A. Meyer wrote:
>>>>>> Hi Fernando
>>>>>>
>>>>>> the next tracking meeting is not until June 23, so if we wait
>>>>>> until then to discuss this
>>>>>> issue, we will certainly fail to meet your deadline of June 25.
>>>>>> It also takes several days
>>>>>> of cosmic running to get a data point, so checking anything also
>>>>>> takes time. If there is any
>>>>>> way that we could have a meeting next week, it would help speed
>>>>>> things along.
>>>>>>
>>>>>> thanks - Curtis
>>>>>>
>>>>>>
>>>>>> On 6/9/10 12:55 PM, Fernando J. Barbosa wrote:
>>>>>>> Hi Naomi,
>>>>>>>
>>>>>>> I suggest we reserve part of the next tracking meeting to
>>>>>>> discuss these issues.
>>>>>>>
>>>>>>> Let me also take the opportunity to emphasize that the CDC and
>>>>>>> the FDC groups must finalize their measurements so that the
>>>>>>> procurement of the ASICs is not impacted. This means that I
>>>>>>> would like to close this project by 25 June 2010.
>>>>>>>
>>>>>>> Thanks and best regards,
>>>>>>> Fernando
>>>>>>>
>>>>>>>
>>>>>>> Naomi Jarvis wrote:
>>>>>>>>
>>>>>>>> Good morning!
>>>>>>>>
>>>>>>>> I would like to talk about the new preamp but I don't think
>>>>>>>> there will be enough time in our tracking mtg this morning.
>>>>>>>> The concerns are that the preamp output is topping out at
>>>>>>>> around 500 mV even with the cosmics, and the gain for cosmics
>>>>>>>> seems to have increased, whereas for Fe55 it was ~1/3 less.
>>>>>>>>
>>>>>>>> I put one page on the wiki
>>>>>>>> http://www.jlab.org/Hall-D/software/wiki/index.php/CDC#Preamp4_.28GlueXdoc1364_Tbl_2_Row_2.29
>>>>>>>>
>>>>>>>> The recent cosmic runs there took a week each, and were with
>>>>>>>> the chamber flat, ie 90 degree cosmics. I'll add more data as
>>>>>>>> it becomes available; at present it is counting at 2kV.
>>>>>>>> Can we set up a time to talk about this next week? (The rest
>>>>>>>> of this week is out for us; I think the only times we cannot do
>>>>>>>> next week are 1pm Tues + the regular mtg on Wed).
>>>>>>>>
>>>>>>>> Thanks,
>>>>>>>>
>>>>>>>> Naomi.
>>>>>>
>>>>>>
>>>>
>>>>
>>
>
> _______________________________________________
> Halld-tracking-hw mailing list
> Halld-tracking-hw at jlab.org
> https://mailman.jlab.org/mailman/listinfo/halld-tracking-hw
--
Prof. Curtis A. Meyer Department of Physics
Phone: (412) 268-2745 Carnegie Mellon University
Fax: (412) 681-0648 Pittsburgh PA 15213-3890
cmeyer at ernest.phys.cmu.edu http://www.curtismeyer.com/