That is great news! Curtis
On 3/23/12 6:56 PM, Lubomir Pentchev wrote:
> I wanted to inform everybody that our last experiment with the
> testing chamber was successful: just coating the O-ring groove in the
> G10 frame with Hysol reduced the oxygen contamination from 1400ppm to
> 300ppm. Taking into account that the incoming gas has 200ppm, the chamber
> now contributes only 100ppm!
>
> The configuration of the testing chamber is, from bottom to top: Lexan
> sheet, plastic spacer ring (with two Buna-N O-rings), empty cathode G10
> frame (with one EPDM O-ring) on which the groove was coated, top Lexan
> sheet.The gas flow is 220scc/min using 40/60 Ar/CO2.
>
> On Monday we will replace the two Buna-N O-rings with the old type EPDM
> ones, to see if there's an effect. In any case we start preparing for the
> chamber refurbishment and will begin with the third package.
>
> Best regards,
> Lubomir