[Halld-tracking-hw] CDC dE/dx and truncation
Naomi Jarvis
nsj at cmu.edu
Tue Aug 15 16:08:10 EDT 2017
The attached files each have a few plots in comparing runs at different rates and changing the dE/dx truncation from 50% to 20% (this is the % of hits which are removed, starting with those with largest dE/dx).
My observations:
The dE/dx from pulse integral broadens at higher rates, quite a bit more than the dE/dx from amplitude does.
Changing the truncation to 20% is very beneficial to the dE/dx from integral, only slightly beneficial to dE/dx from amplitude (the 20% was picked for optimal dE/dx from integral).
dE/dx from amplitude gives better separation than integral, even with 20% truncation. Maybe this would improve further with a different %.
Best,
Naomi.
-------------- next part --------------
A non-text attachment was scrubbed...
Name: CDC_dedx_truncation.pdf
Type: application/pdf
Size: 793619 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/halld-tracking-hw/attachments/20170815/29904584/attachment-0002.pdf>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: CDC_dedx_beamcurrent.pdf
Type: application/pdf
Size: 395221 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/halld-tracking-hw/attachments/20170815/29904584/attachment-0003.pdf>
More information about the Halld-tracking-hw
mailing list