[Hps] ECal high-current preamps resolved
Gabriel CHARLES
charlesg at ipno.in2p3.fr
Thu Apr 9 13:17:37 EDT 2015
Nathan,
As you mention people that have worked on this, I think it is fair to
say that your support (Daq, GUI, software...) has been essential.
Gabriel
Le 2015-04-09 10:50, Nathan Baltzell a écrit :
> Hello Everyone,
>
> The cause of ECal's high-current HV groups has been found to be the
> glue of a label on the
> preamps' circuit boards. Apparently it has become conductive over
> time, resulting in short
> circuiting the HV trace for some preamps. Thanks to Emmanuel for
> diagnosing this with the
> “faulty” preamps we had replaced and sent home.
>
> This week, after only removing the labels, the current on our two
> existing high-current
> channels went down to zero. Thanks to Gabriel and Mathieu, all 442
> labels have been
> relocated and all channels retested with LEDs.
>
> This is important because, over the previous few months, we've seen a
> dozen preamps develop
> this problem and cause their entire group of 10 channels to draw tens
> of uA (some up to 100 uA,
> and vary significantly over time). And since +10 uA results in a ~3%
> gain reduction unless voltages
> are manually adjusted, this is a big improvement for gain stability.
>
> -Nathan
> _______________________________________________
> Hps mailing list
> Hps at jlab.org
> https://mailman.jlab.org/mailman/listinfo/hps