[Moller_intdet] [EXTERNAL] Detector mounting structure
Michael Gericke
Michael.Gericke at umanitoba.ca
Thu Nov 7 23:44:31 EST 2019
Hello everyone,
I am attaching my conceptual design for the detector mounting structure.
Since we haven't had anything of this sort ready for getting meaningful
cost estimates (and I had to do this for the CFI quotes anyway), I decided
to make a design that could be submitted to a company for a quote request.
I am still waiting for them to get back to me.
The attached pdf is CAD generated and you can zoom in and rotate the
models. If you wish to have the CAD files, I'll be happy to upload them
somewhere. The attached movie illustrates how the individual modules
can be assembled (sorry, the files are fairly large, so one has to be
patient).
A few points I want to make:
1) As I said, this is a conceptual design and I am pretty sure it can be
optimized.
I tried to balance mounting rigidity with as small a material
budget as possible,
but I am sure one can further improve on that. Maybe it will
provide some food
for thought for the group working on the structure design.
2) Some of the parts are complicated; The company I spoke to said that
they could
either 3D print or use a 5 axis CNC to make the more complicated
parts, but I expect
the cost to be higher than what we budgeted up to now, especially
if we decided that
we want to us a carbon fiber based material, rather than aluminum.
3) I think I know how one could simplify the complicated parts and make
it cheaper, but at
the cost of less rigidity, I think.
4) This design treats the PMT-LG-Quartz assembly as a unit, which will
make light tight sealing
easier. The mounting plates for each module would allow using
washers to adjust angles, if
need be.
5) The extra space needed by the "skeleton" for each module causes the
rings to be spread out
a little more, in the beam direction. One can play around with the
structure to squeeze that
together, but I would caution against overdoing that, as it could
get impossible to assemble
and service parts. It would be better to adjust quartz tile size
and mounting radii.
6) I think we will need two separate, concentric cylindrical structures,
one that will hold the
modules, as in this design, and a separate one that will hold the
shielding. I don't think we
can afford to have a high weight load on the module structure, as
it was suggested in the
previous designs.
7) We will need to think about whether we need to flush the light guide
with dry air, to prevent
possible reflectivity degradation, which would further complicate
the design. The modules would
support a light and air-tight bag around each one and one could
imagine feeding the air through
the PMT housing.
So, while this still just a preliminary design, I think it can serve as
a first basis for a real cost
estimate of the mounting structure and will show the CDR committee that
we have at least
thought about the mounting details.
As soon as I have a quote for the module parts, I will pass it on to Carl.
Cheers,
Michael
--
--
Michael Gericke (Ph.D., Professor)
Physics and Astronomy
University of Manitoba
30A Sifton Road, 213 Allen Bldg.
Winnipeg, MB R3T 2N2, Canada
-------------- next part --------------
A non-text attachment was scrubbed...
Name: ThinDetectorArrayAssembly.pdf
Type: application/pdf
Size: 13900412 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/moller_intdet/attachments/20191107/92b73b83/attachment-0001.pdf>