[Pansophy] J1.3-09 Document Status for cavity tests

Kirk Davis kdavis at jlab.org
Thu Apr 23 13:51:58 EDT 2020


Valerie,
I recall that when launching an NCR from the VTRF traveler, the SNtype field was never pre-populated with CAVSN. If the tester forgot to type that in, pansophy generated an error. This may have led to losing the actual serial number, and subsequent manual entry error. I recall talking to Meghan about it, but I don’t recall the resolution

Kirk

On Apr 23, 2020, at 1:20 PM, Valerie Bookwalter <bookwalt at jlab.org> wrote:


Anne (et al),

​The P-Team just had a discussion about Naeem's recent finding... NCRs in L2PRD are not matching the CAVSN's of the VTRF travelers.

What we have found, is that in some cases the NCRs were hand types rather auto-populated by the VTRF traveler and thus incorrect numbers are being used. Not only for the CAVSN's but also the SeqNum of the Travelers they should be referring to.

So,.... I was trying to show a list of possible problem, BUT I just lost remote desktop connection. I will finish my investigation and get you the rest of the information as soon as I can connect to JLab again.

This is just a heads up. If anyone finds a disconnect in CAVSNs between VTRF travelers and NCRs, please notify pansophy at jlab.org and we will try to work with the test engineers to resolve the differences.

Valerie / Pansophy Team
________________________________
From: E. Anne McEwen <mcewen at jlab.org>
Sent: Thursday, April 23, 2020 10:10 AM
To: Liang Zhao <lzhao at jlab.org>
Cc: Megan McDonald <megan at jlab.org>; Naeem Huque <huque at jlab.org>; Jacob Harris <jharris at jlab.org>; Valerie Bookwalter <bookwalt at jlab.org>
Subject: Fwd: J1.3-09 Document Status for cavity tests

Hi Liang

Could you please take a look at the open cavity travelers and NCR for VTA on all L2 to add in any missing test informtion

Naeem has been asked to close as much documents as possible for L2 .For the remaining cavities that were not used likely the final disposition will be REJECT. Naeem will be able to go in and close the traveller/NCRs once there is a final disposition and data entered

Please let us know if you have any questions

Anne

Get Outlook for Android<https://aka.ms/ghei36>
________________________________
From: Naeem Huque <huque at jlab.org>
Sent: Wednesday, April 22, 2020 10:06:41 AM
To: E. Anne McEwen <mcewen at jlab.org>
Subject: Re: J1.3-09 Document Status

Thanks Anne, I'll send these out to people
________________________________
From: E. Anne McEwen <mcewen at jlab.org>
Sent: Wednesday, April 22, 2020 10:03 AM
To: Naeem Huque <huque at jlab.org>
Subject: FW: J1.3-09 Document Status




Hi Naeem



Here is the list from J1.3-09 for documents remaining



However now that we hear from Joe that testing is still not done , there is likely more to come



Anne







From: Valerie Bookwalter <bookwalt at jlab.org>
Sent: Wednesday, April 22, 2020 8:20 AM
To: E. Anne McEwen <mcewen at jlab.org>
Cc: Tony Reilly <areilly at jlab.org>; Joe Preble <preble at jlab.org>
Subject: J1.3-09 Document Status



Here is the current status of documents for J1.3-09



J1.3-09R Documentation Status

22Apr2020



Missing data:

From ACTS traveler: still some data to be completed

From SHIP traveler: Not started

Files from SHIP:  all files still to be done

HOM file for cav 348

Open Travelers:

SEQ NUM

TRAV

TRAV TITLE

23

CM-ASSY-SCND

LCLS2 Production Cryomodule Assembly Traveler Cold Mass Phase 2 (WS3)

21

CM-ASSY-VV

LCLS2 Production Cryomodule Vacuum Vessel Assembly Traveler

27

CM-ASSY-FNAL

LCLS2 Production Cryomodule Final Assembly Traveler

17

CM-ACTS-PREP

LCLSII (Production Cryomodule) Install and Cool Down

23

CM-ACTS

LCLS II Production Cryomodule Acceptance Testing

19

CM-INSP-UCM

LCLS-II Upper Cold Mass Incoming Inspection

11

CM-DMAG-UCM

LCLS-II Upper Cold Mass Magnetic Hygiene Control

26

BPM-ASSY

BPM Magnet Sub-assembly Traveler

205

RFIN1

L2PRD LCLS-II Cavity RF Incoming Inspection

204

CAVRFIN7

L2PRD LCLS-II Cavity RF Incoming Inspection

209

CAV-VTA-HOM

LCLSII Cavity Production VTA HOM Survey

234

CAV-VTA-HOM

LCLSII Cavity Production VTA HOM Survey

303

CAVVTRF3

LCLSII (Production Cryomodule) Vertical cavity Testing

319

CAVVTRF3

LCLSII (Production Cryomodule) Vertical cavity Testing

333

CAVCOOL2

LCLS Cavity VTA Cooldown

329

CAVCOOL3

LCLS Cavity VTA Cooldown

Open NCRs:

SEQ NUM

TRAV

DESCRIBE

DISPOSITION

FINAL DISPOSITION

1968

CM-ACTS

Cavity 5 has a FE onset gradient of 13.5 MV/m which is lower than the Minimum Acceptance Criteria.





1967

CM-ACTS

Cavity 4 has a FE onset gradient of 13.5 MV/m which is lower than the Minimum Acceptance Criteria.





1702

CAVINSP4

Tee Header pipe bent.

Hold/Re-measure



1762

CAVINSP7

Distance from cavity beam line to 2 phase header pipe out of tolerance 0.1 mm





1872

CAVVTRF2

FE onset 18 MV/m

Hold/Modify



1473

CAVVTRF2

Emax is below range, 18.5 MV/m

Hold/Re-measure



1876

CAVVTRF3

FE onset 20 MV/m

Hold/Modify



1882

CAVVTRF3

Qo at 16MV/m is 2.3e10, should be 2.5e10 or above.





1835

CAVVTRF3

Final FE onset 19.5 MV/m

Hold/Modify



1443

CAVVTRF3

Cavity quenched at 18.5 MV/m last valid data point recorded was at 18.26 MV/m

Hold/Modify



1627

CAVVTRF5

Blew up field emitter and ended up having FE onset at 11 MV/m

Hold/Modify



1500

CAVVTRF6

Cavity Quench at 15.3 MV/m





1382

CAVVTRF8

Q of field probe was low at 8.61e10





1700

CAVRFIN4

M-Parameter out of spec. Freq. shift in modes plot





1583

CAVRFIN5

M Parameter above 10kHz 38.58 kHz

Proceed to next WorkCenter



1761

CAVRFIN7

M-Parameter larger than 10 kHz measures 14.76 kHz