[Pansophy] Mircea: VTA test results D5 HH-R031/HE-R094/HE-R135 8 May 2024

Mircea Stirbet stirbet at jlab.org
Wed May 15 08:34:17 EDT 2024


Good morning John,

Do not understand why these two travelers have no data... once I have the data processing, I filled up the travelers, and if needed generate the NCR... the NCR is based on Performance limit, I just copy what I have written in this box and paste in the NCR.

This morning, for HE-R135 the performance is empty
[cid:0a5d0789-1d1b-4b7f-bd22-dca0b65a550c]
But.. the NCR 613 has text input on 9 May 2024:
[cid:e2cd787d-cdc1-4564-a709-db7fa98ec3cb]
Something related  with Pansophy ?!

Had , as requested updated data for the two travelers:

Take care,
Mircea

________________________________
From: John Vennekate <hannesv at jlab.org>
Sent: Tuesday, May 14, 2024 14:13
To: Mircea Stirbet <stirbet at jlab.org>
Subject: RE: Mircea: VTA test results D5 HH-R031/HE-R094/HE-R135 8 May 2024


Mircea,



Could you please fill out the corresponding travelers below. I accidentally closed the one for HE-R031 but already asked the Pansophy-team to reopen it for you.





Cheers,

John V

-7607







From: Mircea Stirbet <stirbet at jlab.org>
Sent: Wednesday, May 8, 2024 12:19
To: Kirk Davis <kdavis at jlab.org>; Justin Kent <jtkent at jlab.org>; John Vennekate <hannesv at jlab.org>; Chris Dreyfuss <dreyfuss at jlab.org>; Danny Forehand <forehand at jlab.org>; Alex Wildeson <wildeson at jlab.org>
Cc: Mircea Stirbet <stirbet at jlab.org>; Peter Owen <powen at jlab.org>; Tony Reilly <areilly at jlab.org>; Adam Grabowski <adamg at jlab.org>
Subject: Mircea: VTA test results D5 HH-R031/HE-R094/HE-R135 8 May 2024



Good afternoon,



D5, TS34-002 had low vacuum readings up to the right angle valve (about 2.5e-5 mbar). Usually this value is one or two orders of magnitude higher.

  1.  HE-R031 (Pos1) quench limited at 16.7 MV/m. No FE onset up to Emax.

  1.  HE-R094 (Pos2) radiation onset at low cavity gradient (6-7 MV/m). Qo drop bellow HE project specification at 19.5 MV/m

  1.  HE-R135 (Pos3). In the morning, measured all passbands  frequency  with the Network Analyzer. However, was not able to find the lock frequency during attempted high RF power tests. Suspect lambda leak.



D5 has enough liquid for more testing. This afternoon, Peter will have a look on HE-R135.  Thanks Peter!



See attached screen copies for HE-R031 and HE-R094.



Mircea


-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/pansophy/attachments/20240515/627b9181/attachment-0001.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image.png
Type: image/png
Size: 33923 bytes
Desc: image.png
URL: <https://mailman.jlab.org/pipermail/pansophy/attachments/20240515/627b9181/attachment-0002.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image.png
Type: image/png
Size: 58654 bytes
Desc: image.png
URL: <https://mailman.jlab.org/pipermail/pansophy/attachments/20240515/627b9181/attachment-0003.png>


More information about the Pansophy mailing list