Michelle,
If I change what is pulled from what Tiffany gave me, it will change all of the values for the other cavities. I have attached the spreadsheet from Tiffany but here is what she had me pulling.
Criteria
Traveler
Field
Note
Cavity 3 max rad dose rate
C100R-VTA-CAV-VTRF
Final_RadMax
For the sequence number you mentioned, the Final_RadMax was not filled out
Cavity 3 External-Q of FPC
C100R-TUNE-CAV-RFIN
QextFPC
Not much I can do if there is no traveler
Cavity 8 Emax
C100R-TUNE-CAV-VTRF
Final_EmaxMVm
It looks like there was supposed to be some code to copy the values in that was never put in, I will rectify this
Cavity 8 Qo at Eacc=20MV/m
C100R-VTA-CAV-VTRF
Final_QoAt20MVm
It looks like there was supposed to be some code to copy the values in that was never put in, I will rectify this
Cavity 8 8 Max Rad Dose Rate
C100R-VTA-CAV-VTRF
Final_RadMax
This is the same box as the top criteria here that you gave me a separate field from each other, both of which are different from what Tiffany provided me with.
For the 2 highlighted things, I am hesitant to change the entire report to account for some data that wasn't filled in. Though I am not an expert. The final 3 items would also be populated if the code that is supposed to be in this traveler was there, that is my bad on our end
Megan
________________________________
From: Michelle Oast <weinmann at jlab.org>
Sent: Friday, November 8, 2024 2:35 PM
To: Megan McDonald <megan at jlab.org>
Cc: Ashley Mitchell <ashleya at jlab.org>
Subject: C100-08 Acceptance criteria fixes
Hi Megan,
Here’s the missing data from C100-08’s acceptance criteria page:
Missing Field in CACR
Traveler
Field
Page
Cavity 3 Max rad dose rate
C100R-VTA-CAV-VTRF-97
Final_QoAt20MVm
3
Cavity 3 External-Q of FPC
C100R-TUNE-CAV-RFIN (traveler missing for SN 024)
QextFPC
1
Cavity 8 Emax
C100R-VTA-CAV-VTRF-119
Final_EmaxMVm (left blank, should be 24.15)
3
Cavity 8 Qo at Eacc=20MV/m
C100R-VTA-CAV-VTRF-119
Final_QoAt20MVm (left blank, should be 1.46E+10)
3
Cavity 8 Max rad dose rate
C100R-VTA-CAV-VTRF-119
Final_RadAt20MVm (left blank, should be 1E-3)
3
Let me know if you need anything else, thanks for your help!
Michelle
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/pansophy/attachments/20241111/91b61b26/attachment-0001.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: AcceptCriteria_TravelerList.xlsx
Type: application/vnd.openxmlformats-officedocument.spreadsheetml.sheet
Size: 23234 bytes
Desc: AcceptCriteria_TravelerList.xlsx
URL: <https://mailman.jlab.org/pipermail/pansophy/attachments/20241111/91b61b26/attachment-0001.xlsx>
More information about the Pansophy
mailing list