[Primexd] [New Logentry] CCAL (NPS prototype) beam scan
somov at jlab.org
somov at jlab.org
Sun Dec 2 02:05:01 EST 2018
Logentry Text:
--
We moved each CCAL module to the beam and collected data for calibration and detector resolution studies.
Data were taken for different HV values.
For most CCAL modules, we'll need to set HVs to the relatively small value below 700 V.
Fig. 1 - Fig 5 show the CCAL fadc peak amplitudes as a function of the TAGH counter (energy).
Saturation of FADC amplitudes (the maximum fadc range is 4096 counts) at high voltages
can be seen on Fig. 1 (for 680 v)
Note, we also had to set HVs to the relatively small value during the beam test of the 3x3 prototype,
which we performed with Vlad in Spring 2018, see
https://logbooks.jlab.org/entry/3563789
We are going to check energy linearity using recorded data
(operating CCAL at small voltages may no be a concern, but the gain of
the on-board amplifier (on the divider PCB) should be reduced for the future FCAL II
insert.
We may consider to slightly increased HVs by switching the fadc amplitude
range to 2 V (the current setting is 1 V) after performing CCAL rate studies.
Note, the PMT gain "drops" at smaller voltages for larger anode current
(pulse amplitudes).
----
Fig. 6 FADC peak amplitudes corresponding to the photon beam energy of
about 4.1 GeV (TAGH counters 243) for different HV settings ( CCAL module 4, -5 ).
Dependence of the CCAL amplitude on HV was fit to the power function.
Fig. 7 The same as in Fig. 6 for the CCAL module (2, 2) from the
inner most layer covered by the Tungsten absorber. The beam goes
through a 5 mm absorber hole in the middle of the module.
---