[Rgc] TPOL
Christopher Keith
ckeith at jlab.org
Thu Feb 16 16:35:00 EST 2023
The NMR averaged 36% during Run 17577, but its calibration constant (2.54) is not believed to be very accurate. When we last ran 15 mm proton cells in Sept., we found a higher number (3.67). If we use that, the NMR goes up to 52%. Still lower than the inclusive results, but closer. The online trigger asymmetry (B_DAQ:HEL:60m:29:pol) averaged 69 +/- 0.02%.
Chris
> On Feb 16, 2023, at 3:30 PM, Sebastian Kuhn <kuhn at jlab.org> wrote:
>
> This jibes with the online trigger bit asymmetry-derived Pt, so I find it quite believable. Let’s see if Gregory can confirm (I believe his results are closer to Haruts for NH3 than for ND3).
>
> - Sebastian
>
>> On Feb 16, 2023, at 2:38 PM, Harut Avagyan via Rgc <rgc at jlab.org> wrote:
>>
>> Dear All,
>>
>> Using the same constants, which were typically giving ~10-20% higher Tpols in summer (~80-90%), the new run 17577 gives ~60%, which is still much higher than the NMR
>>
>> https://userweb.jlab.org/~avakian/tmp/all-rgc-run-dep-17574-17578.nh3.pdf
>>
>> More runs cooked will be helpful.
>>
>> Best,
>>
>> Harut
>>
>> _______________________________________________
>> Rgc mailing list
>> Rgc at jlab.org
>> https://mailman.jlab.org/mailman/listinfo/rgc
>