<html>
  <head>
    <meta http-equiv="Content-Type" content="text/html; charset=utf-8">
  </head>
  <body text="#000000" bgcolor="#FFFFFF">
    I have some questions and comments concerning how fast you want to
    "fill" the RICH,<br>
    <br>
    Questions first;<br>
    <br>
    #1 What is the time frame you require for purging the RICH volume
    from initial condition of clean room air, ~ 50% humidity, to your
    final desired number?<br>
    <br>
    #2 What is the maximum final H2O concentration for storage and
    operation?<br>
    <br>
    #3 What is the N2 volume of the detector?<br>
    <br>
    Comments; <br>
    <br>
    N2 gas purge supply details;<br>
    - The N2 purge supply flow is controlled by a mechanical valve
    rotometer, 0-20 slm<br>
    - Purge flow is monitored by a mass flow transducer, 0-20 slm<br>
    - The N2 purge supply valve panel is a pressure system<br>
    <br>
    N2 purge supply flow specifications<br>
    Channel 1 maximum flow - 20 slm (1200 liters per hour/28,800 liters
    per day)<br>
    Channel 2 maximum flow - 20 slm (1200 liters per hour/28,800 liters
    per day)<br>
    <br>
    The above maximum flow specifications remain constant with regard to
    the type or source of the N2 supply.<br>
    <br>
    Thanks,<br>
    George<br>
    <br>
    <div class="moz-cite-prefix">On 10/9/2017 10:19 AM, Marco Mirazita
      wrote:<br>
    </div>
    <blockquote type="cite"
      cite="mid:cd6690bbc2cb8b4af65fa50c9a3f7e3e@webmail.lnf.infn.it"><br>
      <br>
      - Nitrogen
      <br>
      1) Setup the N2 distribution lines inside the RICH. <br>
      2) The dewar is not suitable to fill the RICH quickly (i.e. in few
      hours). A new system must be used, for example bottles. This
      should be discussed timely with the DSG. There is the plan to test
      the N2 system in the RICH for the week of October 16. This test
      will include installation of the exit panel, rotation of the RICH
      in vertical position, installation of the frontal panels (possibly
      using the assembly frame), sealing of the electronic panel region
      with plastic. It makes sense to do all this assembly work only if
      we have (or we may have) a way to fill the RICH with N2 in a
      reasonable time.
      <br>
      <br>
      <br>
      <br>
      <br>
      _______________________________________________
      <br>
      Dsg-rich mailing list
      <br>
      <a class="moz-txt-link-abbreviated" href="mailto:Dsg-rich@jlab.org">Dsg-rich@jlab.org</a>
      <br>
      <a class="moz-txt-link-freetext" href="https://mailman.jlab.org/mailman/listinfo/dsg-rich">https://mailman.jlab.org/mailman/listinfo/dsg-rich</a>
      <br>
    </blockquote>
    <br>
    <pre class="moz-signature" cols="72">-- 
George Jacobs
Jefferson Lab (TJNAF)
STE 12
12000 Jefferson Ave.
Newport News, VA 23606

(office) 757-269-7115

(cell)   757-876-0480

(email) <a class="moz-txt-link-abbreviated" href="mailto:jacobsg@jlab.org">jacobsg@jlab.org</a>

(website) <a class="moz-txt-link-freetext" href="https://userweb.jlab.org/~jacobsg">https://userweb.jlab.org/~jacobsg</a></pre>
  </body>
</html>