[Sane-analysis] Bigcal Re-Calibration Progresss
Whitney R. Armstrong
whit at temple.edu
Fri Jun 13 13:06:03 EDT 2014
Hello Everyone,
I wanted to give a brief update regarding the Bigcal re-calibration and
summarize the conclusions from our last meeting.
First, I showed on Wednesday that the new calibration *significantly* impacts
the asymmetry results. The new calibrations improve the overall behavior bin to
bin and are consistent with A_parallel measured by CLAS.
Secondly, when looking at the asymmetries for the RCS and Protino separately
(http://quarks.temple.edu/~whit/SANE/analysis_main/results/combined-split_asym_para59_251.png)
it was noted that between x ~ 0.3 to 0.4 there is clear difference between the
top and bottom. Possible explanations include the calibrations, noisy/bad
blocks, or background contamination (through differing kinematics). The latter
is nearly ruled out because it persists through the background subtraction (as
seen here
http://quarks.temple.edu/~whit/SANE/analysis_main/results/asymmetries/A1A2_59_noel_split_253.png).
However, it is possibly a combination of a noisy block causing bad causing the
average kinematics to be skewed and thus calculating the wrong background
correction. For reference here is the background corrections plot
http://quarks.temple.edu/~whit/SANE/analysis_main/results/asymmetries/bg_corrections_para59_251.png.
Or it is the calibration...
In order to provide a check, I have been looking at the eta mass peak (548
MeV/c^2) in the two photon mass spectrum. After spending two days iterationg
over just the two sections of bigcal to improve the calibrations, I produce the
following fits.
The mass peak in the perpendicular calibration looks good (m_0 ~ 550):
http://quarks.temple.edu/~whit/SANE/analysis_main/results/pi0calibration/FitEtaMass_perp_517.png
For the parallel calibration, the mass peak is shifted to lower values by about
20-30 MeV (m_0 ~ 520).
http://quarks.temple.edu/~whit/SANE/analysis_main/results/pi0calibration/FitEtaMass_para_517.png
Previously I had reported that the perpendicular calibration seemed to need
more work. This was true when looking at the overall shift of the pi0 mass
peak, but, when looking at the latest changes in the calibration coefficients
for the parallel they are quite large (bottom right plot
http://quarks.temple.edu/~whit/SANE/analysis_main/results/pi0calibration/pi0CalibrationCompare_para_517_c1_0.png).
Because the changes are large and fluctuate around zero a shift in the pi0mass
peak is not observed (just a narrowing is), but these changes are just as
important as the perpendicular.
I will continue to try to improve the calibrations by focusing on smaller
sections (instead of all of RCS or Protvino at a time). My hope is that with
improvement the parallel calibration's eta mass peak will move towards 550. It
appears to be doing so rather slowly. Look at these plots in order:
http://quarks.temple.edu/~whit/SANE/analysis_main/results/pi0calibration/FitEtaMass_para_514.png
http://quarks.temple.edu/~whit/SANE/analysis_main/results/pi0calibration/FitEtaMass_para_515.png
http://quarks.temple.edu/~whit/SANE/analysis_main/results/pi0calibration/FitEtaMass_para_516.png
http://quarks.temple.edu/~whit/SANE/analysis_main/results/pi0calibration/FitEtaMass_para_517.png
Thank you for your time, and as always, your comments and questions are
appreciated.
Cheers,
Whit
On Wed, Jun 04, 2014 at 03:29:54PM -0400, O. A. Rondon wrote:
>Hi Whit,
>
>Neat results. I agree that for perp the eta might be there, but I would
>suggest plotting errors on the histos (both 80 and 180 data) to see if
>at 80 deg it's not just statistics, like the big jump near 300 (MeV)
>probably is.
>
>Could you also add an explanation of what is being plotted in each
>panel? The meaning of the mass plot is clear (at least of the blue
>histo), and I guess the mass vs xi, yi (i= 1, 2) probably are the mass
>distributions vs the x, y coordinates of clusters 1 and 2 used to
>reconstruct the eta, but I'm not sure about the red mass histo in the
>mass 3 panel; the E2 panel's red and black histos; what D is in the
>massVSD; and what the clustx vs y1, y2 are (the y1,y2 scales here don't
>agree with the scales of the mass vs y1, y2 plots).
>
>Before devoting too much effort in finding the eta in the perp data, I
>would suggest recalculating A180 with the new calibration, to see if it
>makes a difference in the asymmetry.
>
>Cheers,
>
>Oscar
>
>Whitney R. Armstrong wrote:
>> Hello Everyone,
>>
>> I wanted to share some results in my refinement of the BigCal calibration from
>> pi0s.
>>
>> Below you will find plots showing the mass spectrum (above the pi0 mass) from
>> two photons (top central plots) for parallel and perpendicular configurations.
>>
>> In the mass plot you can see a bump near the eta mass (547 MeV).
>>
>> Parallel:
>> http://quarks.temple.edu/~whit/SANE/analysis_main/results/pi0calibration/etamass2-2000_1.png
>>
>> Perp:
>> http://quarks.temple.edu/~whit/SANE/analysis_main/results/pi0calibration/etamass2-2000_5.png
>>
>> Please note the following:
>>
>> 1) Only after recalibrating did the eta mass peak start to show up. This is
>> because the mass reconstructino of the eta is most sensitive to the energy
>> resolution and less sensitive to the position (due to the larger opening angle
>> for the eta compared to the pi0).
>>
>> 2) The parallel calibration is better than the perpendicular. That is, the perp
>> calibration can certainly be improved compared to the parallel, thus the eta
>> mass is resolved better for the parallel.
>>
>> 3) The background is much higher than the pi0, therefore, it does not seem
>> feasible to directly calibrate (like pi0) using the eta mass peak. However,
>> because the two eta clusters are separated bny large distances (>90 cm), it
>> provides a nice confirmation that the opposite sides (eg the top and bottom)
>> are calibrated properly against each other.
>>
>> 4) All previous results were contaminated by the insidious behaviour of the old
>> calibration. The top (RCS) energies were shifted higher while the bottom
>> (Protvino) was shifted lower (thus producing a nicely centered pi0mass when
>> summing over the whole detector). This likely exlains why Anusha consistently
>> saw energy differences in the elastics. It also explains (the problem seen long
>> ago) that the top appeared to have all the higher energy events compared to the
>> bottom. It would be interesting to look at both of these problems again.
>>
>> Please let me know if you have any suggestions on how to clean up the eta mass
>> spectrum.
>>
>> As always, your comments and questions are greatly appreciated.
>>
>> Cheers,
>> Whit
>>
>>
>> _______________________________________________
>> Sane-analysis mailing list
>> Sane-analysis at jlab.org
>> https://mailman.jlab.org/mailman/listinfo/sane-analysis
>>
>
>
>
>Also,
>
>_______________________________________________
>Sane-analysis mailing list
>Sane-analysis at jlab.org
>https://mailman.jlab.org/mailman/listinfo/sane-analysis
More information about the Sane-analysis
mailing list