Thank Ezekiel!
Most likely J0-Bot GEM is fine; discussing the lower HV current with Fausto and Roberto, we ended up with two potential explanations:
1) some divider components (resistors and/or capacitors) got broken or soldering deteriorated; this occurred couple of times in the past
2) the HV power supply channel reading circuitry has damaged; this also appended at least one time, in different circumstances
Possible actions to try to identify the source of problem:
- change HV power supply channel
- replace the divider (this is probably not trivial in the current setup; maybe it is a bit easier to unplug the divider from GEM, test the current vs HV again, if current is still lower then expected, likely the divider need fixing
---