[Sbs_gems] [New Logentry] Preliminary efficiency results from UV layer 1

jeffas at jlab.org jeffas at jlab.org
Tue Jan 11 08:45:01 EST 2022


Logentry Text:
--
The table below is the results from the efficiency scan of UV layer 1. The efficiencies are all lower than we expected, but the timing distribution for this module seems broader than the others, https://logbooks.jlab.org/entry/3969801, so there may be something going on. We see if we look at the same run, but remove the timing cuts the efficiency jumps up by 10%, https://logbooks.jlab.org/entry/3969830. So right now I am skeptical of the absolute value numbers and suspect they can be improved in analysis. Hopefully Andrew has some ideas here. 
 Regardless the shape of the curve shows a plateau starting around 3628 V, so we will set our voltage there for now.


UVa UV Layer 1 / Module 1			
Voltage	Run 	Efficiency	     Error
3653	13184	72.29	     0.64
3628	13186	71.52	     0.66
3608	13187	67.2	             0.67
3564	13188	62.17	     0.73
3519	13189	50.26	     0.76
3475	13190	38.86	     0.71
3430	13191	28.23	     0.68
3385	13192	17.00	     0.56

---