[Sbs_gems] [New Logentry] Issues with GEM HV mpod crate ISEG modules
adr at jlab.org
adr at jlab.org
Mon Feb 20 18:20:03 EST 2023
Logentry Text:
--
The HV channels "inline_uva_l2m0" and "linline_l2m1" started working fine after I rebooted the mpod HV crate. However, then I saw that channels "inline_uva_l2m2" through "inline_uva_l4_m1" would not come up. These channels are connected to one single ISEG HV module. When I power cycled the crate, the set voltages got wiped away and trip thresholds had set to 1000 uA. It was same after a couple of resets and then I decided the whole ISEG HV module has gone bad so I swapped it out with a spare module that happened to be in the crate. The module that had gone bad seems to be of a different version than the reset of the modules we are using right now.
After installing the new ISEG HV module, all the 8 HV channels come up fine and holding HV. See the attached before and after screenshots of the epics GUI.
---
This is a plain text email for clients that cannot display HTML. The full logentry can be found online at https://logbooks.jlab.org/entry/4135337
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/sbs_gems/attachments/20230220/222a7fc9/attachment.html>
More information about the Sbs_gems
mailing list