[BTeam] NL gradient calibration results

Jay Benesch benesch at jlab.org
Tue Nov 28 10:29:30 EST 2023


Colleagues,

The spreadsheet attached has the results of the rfcal run taken of the 
entire NL in order over ten hours with 1E-3 energy lock offset and 
comparison with data taken with somewhat different conditions earlier 
that day and the previous day - see comments column.

The 11/16 data beginning 1545 was taken with
IPM1A04 2.51 m etax
IPM1A21 7.59 m etax
IPM1A26 -2.39 m etax
either disabled or excluded from BEM

The earlier data was taken with IPM1A26 included.

All of the SL data was taken with all BPMs active.  The sum of the 
calibrated momentum gains was 1037.3 MeV vs 1047 MeV per the arc.

In the NL, without these three BPMs, the sum is 1180 MeV, 12.7% above 
1047.  I conclude that BEM's calculation and therefore the energy lock 
is not accurate without at least 1A21 and perhaps all three BPMs.  This 
may explain why the polarization has been off expected values in all 
halls and why HALLC:p is 20 MeV below that expected from eDT.

Jay