[d2n-analysis-talk] mwdc 4-pass t0
    posik at jlab.org 
    posik at jlab.org
       
    Wed May  5 13:54:39 EDT 2010
    
    
  
Hi all,
 So I adjusted the t0 for each plane in the mwdc. I then checked the
residuals for each plane. Before the calibration the residual sigma for
all planes was between ~600-800 um and now they take on values of
~200-300 um. Some planes are better than others. Below I show a plot of
the v3p planes t0's and residuals before and after the t0 adjustment.
  To get an idea of where we are compared to transversity, I talked to Xin
and for his elastic runs he had a resolution of 180-200 um and for
production run he had 200-250um. So we differ only by ~50um.
I am wondering if it is worth doing the calibration wire by wire for the
t0,wire offsets drift distance to drift time conversion?
-matt
-------------- next part --------------
A non-text attachment was scrubbed...
Name: t0.png
Type: image/png
Size: 43873 bytes
Desc: not available
Url : https://mailman.jlab.org/pipermail/d2n-analysis-talk/attachments/20100505/6b37daa7/attachment-0002.png 
-------------- next part --------------
A non-text attachment was scrubbed...
Name: residual.png
Type: image/png
Size: 39688 bytes
Desc: not available
Url : https://mailman.jlab.org/pipermail/d2n-analysis-talk/attachments/20100505/6b37daa7/attachment-0003.png 
    
    
More information about the d2n-analysis-talk
mailing list