marc,
make the plot from 1200 onwards i.e. for flow 150 -250 lpm
it appears that at these  flow rates the temperature of RTD C is around  27 C


________________________________
From: Dsg-eic_tracker <dsg-eic_tracker-bounces at jlab.org> on behalf of Marc Mcmullen via Dsg-eic_tracker <dsg-eic_tracker at jlab.org>
Sent: Thursday, May 4, 2023 9:15 AM
To: Marc Mcmullen via Dsg-eic_tracker <dsg-eic_tracker at jlab.org>
Cc: Patrizia Rossi <rossi at jlab.org>
Subject: [Dsg-eic_tracker] EIC beam line test stand flow test

Hi Patrizia and Rolf,

The flow tests are complete. The beam line pipe was heated to ~100C and air was flowed at 10, 20, 50, 100, 150, 200, and 250 liters per minute. Each flow test ran for 3 hours. After each test the controls waited for 3 hours to allow monitored temperatures to recover to approximately the same starting temperature for every test.

Sensor locations:

RTD_A, B, and C - inner surface of outer pipe (silicon)

RTD_D - outer surface of inner pipe (beryllium beam pipe)

RTD_H1, and H2 - outer surface of the heated pipe (for controls)

Brian,

The full data log is located at O:\DSG\EIC\Software\LabVIEW\Teststand_data\0502_0504_flowtest.xlsx<file:///O:\DSG\EIC\Software\LabVIEW\Teststand_data\0502_0504_flowtest.xlsx>


Regards,
Marc