[Dsg-eic_tracker] EIC beam line test stand flow test

Marc Mcmullen mcmullen at jlab.org
Thu May 4 09:15:46 EDT 2023


Hi Patrizia and Rolf,

The flow tests are complete. The beam line pipe was heated to ~100C and air was flowed at 10, 20, 50, 100, 150, 200, and 250 liters per minute. Each flow test ran for 3 hours. After each test the controls waited for 3 hours to allow monitored temperatures to recover to approximately the same starting temperature for every test.

Sensor locations:

RTD_A, B, and C - inner surface of outer pipe (silicon)

RTD_D - outer surface of inner pipe (beryllium beam pipe)

RTD_H1, and H2 - outer surface of the heated pipe (for controls)

Brian,

The full data log is located at O:\DSG\EIC\Software\LabVIEW\Teststand_data\0502_0504_flowtest.xlsx<file:///O:\DSG\EIC\Software\LabVIEW\Teststand_data\0502_0504_flowtest.xlsx>


Regards,
Marc
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-eic_tracker/attachments/20230504/3628da6e/attachment-0001.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: 0502_0504_flow test plot.jpg
Type: image/jpeg
Size: 206196 bytes
Desc: 0502_0504_flow test plot.jpg
URL: <https://mailman.jlab.org/pipermail/dsg-eic_tracker/attachments/20230504/3628da6e/attachment-0001.jpg>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: EIC beampipe temp_05023_end of flow test.JPG
Type: image/jpeg
Size: 163086 bytes
Desc: EIC beampipe temp_05023_end of flow test.JPG
URL: <https://mailman.jlab.org/pipermail/dsg-eic_tracker/attachments/20230504/3628da6e/attachment-0001.jpe>


More information about the Dsg-eic_tracker mailing list