[Dsg-eic_tracker] EIC simulated beam line flow tests
Marc Mcmullen
mcmullen at jlab.org
Sat May 13 08:13:16 EDT 2023
Hi Patrizia and Rolf,
After discussing the matter with the EHS group (fire protection engineer, industrial hygiene, and pressure systems design authority), we determined that we could safely run the test stand at a higher temperature as long as we keep the safety interlock set well below the auto-ignition temperature of the mineral oil (300 C).
We increased the temperature of the test stand to around 210 C and made adjustments to the airflow to reach 100 C (RTD_D) on the beam pipe while cooling the inner surface of the outer pipe (silicon region) to ~30 C (RTD_A, B, and C).
Attached is the plot from yesterday's test. It looks like setting to heater to 210 C, with a flow of about 212 - 215 Lpm gets us right to those temperatures.
Have a great weekend,
Marc
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-eic_tracker/attachments/20230513/5da7a2db/attachment-0001.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: EIC beampipe temp_051123_flowtest.JPG
Type: image/jpeg
Size: 162643 bytes
Desc: EIC beampipe temp_051123_flowtest.JPG
URL: <https://mailman.jlab.org/pipermail/dsg-eic_tracker/attachments/20230513/5da7a2db/attachment-0001.jpe>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: 0512 EIC beam line flow test_100 pipe_30 si_corrected.jpg
Type: image/jpeg
Size: 213142 bytes
Desc: 0512 EIC beam line flow test_100 pipe_30 si_corrected.jpg
URL: <https://mailman.jlab.org/pipermail/dsg-eic_tracker/attachments/20230513/5da7a2db/attachment-0001.jpg>
More information about the Dsg-eic_tracker
mailing list