[Dsg-halla_ecal] [New Logentry] Heaters restarted more fans installed all setpoints 290degC.

adaq at jlab.org adaq at jlab.org
Fri Apr 18 12:20:01 EDT 2025


Logentry Text:
--
The heaters all turned off around 10:30AM. I went into the hall and found that one Omega for the front ambient temperatures was tripped (furthest left on the rack which I believe is monitoring the bottom front ambient temperature of the detector). All Omega process controllers had sepoints of 270degC and were far from reaching that so perhaps it was a transient or poor thermocouple connection. If it happens again, we will have to try to figure out what is causing it. I reset it and changed the setpoints on all three to 290degC. Temperatures have started to climb again.<br>


---