Hello Marc,

 I concur with the selection but why is 1500ft of cable needed?
Aren't there only two connectors on the CCR for the temperatures?

Steven

________________________________
From: Dsg-halla_magnets <dsg-halla_magnets-bounces at jlab.org> on behalf of Marc Mcmullen <mcmullen at jlab.org>
Sent: Friday, May 14, 2021 8:51 AM
To: dsg-halla_magnets at jlab.org <dsg-halla_magnets at jlab.org>
Subject: [Dsg-halla_magnets] CCR Temperature sensor cable selection


Hi Steve,



DSG has concluded research on cables for the CCR temperature sensors. I have attached the SoLID cable selection document. The CCR temperature sensor cables section on the bottom. All three cables meet the requirements listed below:



32 conductors or greater

Rated for 300V

22 – 24 awg

FT1 rated or better





DSG recommends #3, it is the Olympic version of #2 (Belden). Let me know if you concur with this.



Regards,

Marc McMullen

Sr. Instrumentation Designer

Detector Support Group

(757) 269-7738