[Dsg-halla_magnets] CLEO
Pablo Campero Rojas
campero at jlab.org
Mon Jul 18 10:08:44 EDT 2022
Hello Steven,
Last week I checked the diode sensors wiring from the connector at the CCR to the PLC ADC module and I did not find anything wrong.
I started looking to diode temperature sensor (named TD_He_Mag_Sup) and I noticed a high voltage read-back (measured ~ 6.3 V), which is out of the set/expected voltage range. Based on the Curve-10 for the diodes, the expected voltage at ambient temperature is ~ 0.53 V.
I disconnected the Inst-A connector at the CCR side and I plugged a dummy resistor of 1 KOhms in the connector and measured the voltage in the control rack terminal blocks (measured ~0.01 V). I made sure the excitation current was correct (measured the expected 10uA coming out from the CCS board).
Did you test the diode resistance at ambient temperature before?
Please confirm if the attached specs sheet corresponds to the diode sensors installed inside the CCR?
Today I'm working remotely on the FTView data logging model and making some changes to the Radial & Axial support expert screen.
I will at the lab tomorrow if you're available we can meet and go over these topics.
--Pablo
________________________________
From: Steven Lassiter <lassiter at jlab.org>
Sent: Monday, July 18, 2022 9:09 AM
To: Pablo Campero Rojas <campero at jlab.org>
Subject: CLEO
Hello Pablo,
Will you be working on CLEO signals any today?
Also, would like to help on setting up and running the CLEO HMI on other computers. Seems I do not have access to some files.
Thanks
Steven
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-halla_magnets/attachments/20220718/441e405b/attachment-0001.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: lstc_dt400_l- Diode temp sensors.pdf
Type: application/pdf
Size: 1391952 bytes
Desc: lstc_dt400_l- Diode temp sensors.pdf
URL: <https://mailman.jlab.org/pipermail/dsg-halla_magnets/attachments/20220718/441e405b/attachment-0001.pdf>
More information about the Dsg-halla_magnets
mailing list