[Dsg-dc] DC Gas Mixing
Marc McMullen
mcmullen at jlab.org
Thu Oct 11 16:37:03 EDT 2018
DSG held a training session with Hall B Mechanical this morning. The
topic was calibrating the DC gas mixture to the 10% CO2/ 90% Ar premixed
standard.
The premixed gas standard is used as the direct reference for the
current value (in mA) of the gas mixture; 10% CO2/ 90% Ar.
Today's training was done only using Mix 1, as it has the only working
thermal conductivity gas analyzer. All measurements were taken using the
Mix 1 analyzer.
The gas standard and mix values are very close (~8.4mA, ~2.74V),
however, the sample coming from the buffer tank (R1 MFC output)
indicates a higher CO2%.
Monday Hall B Mechanical will be able to connect the Mix 2 gas sample
line to the manifold which is connected to Mix 1 analyzer. At that time
we will isolate and zero all mixing MFCs and run the calibration tests
on both mixing systems. After both mixing systems have been calibrated
Hall B management should consider purchasing a replacement for the CLAS
6 (purchased mid 90s, last calibrated in 2000) analyzers for both mixing
units (starting at $3700 per unit), and additional units for the output
of the tanks.
The results of the analysis were as follows:
--
Marc McMullen
Detector Support Group
757-269-7738
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-dc/attachments/20181011/b2013139/attachment-0001.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: 101118 TCU Results.JPG
Type: image/jpeg
Size: 58145 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/dsg-dc/attachments/20181011/b2013139/attachment-0001.jpe>
More information about the Dsg-dc
mailing list