[Dsg-dc] DC Gas Mixing
Marc McMullen
mcmullen at jlab.org
Thu Oct 11 16:49:50 EDT 2018
I had an incomplete sentence;
After both mixing systems have been calibrated, we will observe the
mixture and buffer tank CO2% levels to see if it changes. This will take
up to 5 volume exchanges to see any effect.
On 10/11/2018 4:37 PM, Marc McMullen wrote:
>
> DSG held a training session with Hall B Mechanical this morning. The
> topic was calibrating the DC gas mixture to the 10% CO2/ 90% Ar
> premixed standard.
>
> The premixed gas standard is used as the direct reference for the
> current value (in mA) of the gas mixture; 10% CO2/ 90% Ar.
>
> Today's training was done only using Mix 1, as it has the only working
> thermal conductivity gas analyzer. All measurements were taken using
> the Mix 1 analyzer.
>
> The gas standard and mix values are very close (~8.4mA, ~2.74V),
> however, the sample coming from the buffer tank (R1 MFC output)
> indicates a higher CO2%.
>
> Monday Hall B Mechanical will be able to connect the Mix 2 gas sample
> line to the manifold which is connected to Mix 1 analyzer. At that
> time we will isolate and zero all mixing MFCs and run the calibration
> tests on both mixing systems. After both mixing systems have been
> calibrated
>
> Hall B management should consider purchasing a replacement for the
> CLAS 6 (purchased mid 90s, last calibrated in 2000) analyzers for both
> mixing units (starting at $3700 per unit), and additional units for
> the output of the tanks.
>
> The results of the analysis were as follows:
>
> --
> Marc McMullen
> Detector Support Group
> 757-269-7738
>
>
> _______________________________________________
> Dsg-dc mailing list
> Dsg-dc at jlab.org
> https://mailman.jlab.org/mailman/listinfo/dsg-dc
--
Marc McMullen
Detector Support Group
757-269-7738
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_dc/attachments/20181011/947fe0bb/attachment-0002.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: 101118 TCU Results.JPG
Type: image/jpeg
Size: 58145 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_dc/attachments/20181011/947fe0bb/attachment-0002.jpe>
More information about the Dsg-dc