[Dsg-ft] Interlock test
Raffaella De Vita
Raffaella.Devita at ge.infn.it
Mon May 15 12:29:08 EDT 2017
Hi Amanda,
as I mentioned on Friday, we disconnected HV and LV cables so that you
can safely test the interlocks without worrying about creating some
unusual condition that could affect the detector. It should therefore be
safe for you to proceed.
For LV cable, I believe what will happen is that, with the cable
disconnected from the detector, the sense wires are not properly
attached and the board will still output voltage but read-backs will be
wrong. I just got back to Genova and I don't have the information with
me but should be able to tell you which pins should be shorted tomorrow
when I will be in the office.
Regards,
Raffaella
Amanda Hoebel wrote:
> Hi Raffaella,
>
> We are about to start the hardware interlock tests and would like to know the status of the detector. We noticed the HV and LV have been disconnected from the calorimeter. Is it ok to start our tests now? Also what pins should be shorted on the LV cable so that the LV does not trip?
>
> Thanks,
> Amanda