Dear All,
you can find in attachment the LV cable diagram. To ensure the voltage 
sense function will work properly and be able to test the interlocks 
with the cable disconnected, you should:
- short pin 1 and 2
- short pin 4 and 5
on connectors J2 and J3 (white 6 pin connectors).
Note that the pin numbering shown for J2 and J3 in the diagram is what 
you would see looking at the connector from the wires side.
Let us know if you have any question.
Regards,
     Raffaella

Raffaella De Vita wrote:
> Hi Amanda,
> as I mentioned on Friday, we disconnected HV and LV cables so that you 
> can safely test the interlocks without worrying about creating some 
> unusual condition that could affect the detector. It should therefore 
> be safe for you to proceed.
> For LV cable, I believe what will happen is that, with the cable 
> disconnected from the detector, the sense wires are not properly 
> attached and the board will still output voltage but read-backs will 
> be wrong. I just got back to Genova and I don't have the information 
> with me but should be able to tell you which pins should be shorted 
> tomorrow when I will be in the office.
> Regards,
>     Raffaella
>
> Amanda Hoebel wrote:
>> Hi Raffaella,
>>
>> We are about to start the hardware interlock tests and would like to 
>> know the status of the detector. We noticed the HV and LV have been 
>> disconnected from the calorimeter. Is it ok to start our tests now? 
>> Also what pins should be shorted on the LV cable so that the LV does 
>> not trip?
>>
>> Thanks,
>> Amanda
>> _______________________________________________
>> Dsg-ft mailing list
>> Dsg-ft at jlab.org
>> https://mailman.jlab.org/mailman/listinfo/dsg-ft
>

-------------- next part --------------
A non-text attachment was scrubbed...
Name: Cablaggio LV.pdf
Type: application/x-download
Size: 9124 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_ft/attachments/20170516/a8dbe8f2/attachment-0002.bin>