[Dsg-ft] Addition of condensation sensor to FT Hardware Interlock System

Raffaella De Vita Raffaella.Devita at ge.infn.it
Mon Nov 27 16:56:55 EST 2017


Hi Pete,
thanks for sending the updated chart. The diagram for the new sensor 
looks good to me. Would it be possible to increase the delay between the 
HV turnoff and the LV turnoff from 2 to 10 sec? this would also apply to 
the humidity and temperature interlocks. I noticed in fact recently that 
the HV ramp down takes more than the 2 sec with the settings we have on 
the caen boards at the moment.
Thanks,
     Raffaella

Peter Bonneau wrote:
>
> A condensation sensor is being added to the FT Hardware Interlock 
> System. The attached fault chart for the calorimeter has been updated 
> for the addition of the condensation sensor.
>
> If condensation is detected, the interlock system will :
>
> 1) Disable Chiller
>
> 2) Disable calorimeter HV
>
> 3) Wait ~ 2 Seconds
>
> 4) Disable calorimeter LV.
>
>
> To support the new sensor, the LabVIEW and EPICS user interfaces will 
> have condensation interlock enable/disable controls, signal status, 
> and latched interlock indicators. In addition, the new signals will be 
> added to the Mya datalogger.
>
>
>
>
>