[Dsg-ft] Discussion about thermal insulation of MFC
Marc McMullen
mcmullen at jlab.org
Wed Jan 3 10:36:00 EST 2018
Hi All,
As an update, we are expecting 10 bottles of premix around the 3 or 4th
week of January. If it comes early, I'll notify the group email. We have
also order another 20 bottles, which should bring the total premix to 30
bottles (e.g. ~90 running days).
Marc
On 1/2/2018 8:34 PM, mdefurne wrote:
> Hi Bob, Hi Todd,
>
> First of all I wish you all the best for this new year 2018. And thank
> you Todd for installing the CF4 bottle.
>
> Tomorrow I would like to discuss with you about the insulation of the
> mixing system. For instance, the metallic plate of the MFCs is
> directly in contact with the metallic wall of the system. Don t you
> think the metallic walls may act as a heat sink by conduction?
>
> Also I am wondering why we cannot close the entire system with wood
> panels and put a heater inside that will warm the air inside. Keeping
> the air at 70F would be a lot more smoother and efficient to keep the
> MFCs at good temperature than turning off-on the heat tape on the
> MFC-plate. I am not sure that heat tape is the most efficient way to
> keep their temperature high enough and maybe most importantly stable.
> Indeed the heat tape will be turned on and off depending on the
> outdoors temperature at a more or less relatively high-rate (You can
> hear the thermostat turning on and off the heat tape every 5 seconds
> currently). Will it not prevent the MFC from performing smoothly?
> Because we need also to have a stable mixture of gas. Changing the
> isobutane fraction by 1% makes a significant difference in the
> detector performances.
>
> Moreover, even mixed, the isobutane may condensate. It is of course
> much more complicated to have it condensated than when it is pure
> (colder temperature and no flow). But I am willing to bet that it has
> condensated in the small tank of the barrel with some premix since the
> temperature has been really really and unusually cold and we did not
> draw gas for 10 days. We will wait for monday, when temperature will
> be much higher to flush everything with Argon to make sure there is no
> inhomogeneities of mixture due to this condensation episode.
> Nevertheless, we should try to prevent the isobutane from condensating
> anywhere for the detector safety. This complete enclosure thing would
> also keep the small tanks warm enough. Moreover I think it is a
> relatively cheap and sure solution.
>
> After discussing with Marc and Brian, we agreed that turning the
> mixing system ON should be interlocked with all heating systems
> related to keeping MFCs or isobutane warm enough. If a heat tape or
> heater fails, we should not take the risk to break a MFC.
>
> Finally, even if we plan to have the MFC repaired, I think we should
> buy a few MFCs (2 or 3) as extra-spares. Delivery time is long (4-6
> weeks), and we have no guarantee that everything will work smoothly
> when we turn the system back on. Not as MVT expert but more as
> spokesperson of the next experiment, we must have the MVT working. The
> SVT has been damaged as you know and is performing badly compared to
> what was expected. For the central tracking, MVT is now absolutely
> necessary. I would not recommend to try saving money on 2 or 3 extra
> MFCs. It will still be way cheaper than running with premix bottles
> for 4-6 weeks while we are still trying to debug the system.
>
> I have a meeting from 8:30 to 10:30 so I can come to the Hall around
> 11am.
>
> Best regards,
>
> Maxime
>
>
--
Marc McMullen
Detector Support Group
757-269-7738
More information about the Dsg-ft
mailing list