Hello Ruben, 

Tyler, Amanda and I tested the following instrumentation: 
1. PT-100 temperature sensor TP8675 (Vaporizer) 
2. PT-100 temperature sensor TP8680 (Relief Valve) 
3. Cryocon Units 2 and 3. 
4. Solenoid valve SV8622 (Vent valve) 

Controls for the temperature sensors and solenoid valve are working as expected, and also readouts of these instrumentation was checked. 
Detailed information of the results for this test can be find in the TEST PLC- EPICS PLAN_v7_June 16th 2017 spreadsheet. Find it on M:\hallb_eng\CLAS12\Magnets\Solenoid\Controls and Wiring. 


Wesley, 
During the test, the 'Solenoid Cryocon Status' EPICS screen was checked too, and we noticed the following: 
1. 'Solenoid Watchdog' indicator is not updating for Solenoid and DBX. (pink color in the Solenoid-Watchdog readout box) 
2. ' Reset Button' is not connecting with the associate tag in the Solenoid PLC code. 

Here is the tag names used in the Solenoid PLC code that are related with these signals mentioned above. 

PLC TAG 	Data Type 
Sol_CryoCon_Reset_PB 	BOOL 
Sol_CryoCon_Watchdog 	INT 

If you have a chance, please check and let me know any comments. 


Best Regards, 
Pablo 


From: "Ruben Fair" <rfair at jlab.org> 
To: "Pablo Campero Rojas" <campero at jlab.org> 
Cc: "dsg-hallb magnets" <dsg-hallb_magnets at jlab.org> 
Sent: Tuesday, June 13, 2017 8:58:13 AM 
Subject: Re: Solenoid Controls_Final_Wiring check-out 

Thank you very much Pablo - appreciate it. 

Let me know if you need help. 

Cheers 

Ruben 



From: "Pablo Campero Rojas" <campero at jlab.org> 
To: "Ruben Fair" <rfair at jlab.org> 
Cc: "dsg-hallb magnets" <dsg-hallb_magnets at jlab.org> 
Sent: Tuesday, June 13, 2017 8:56:19 AM 
Subject: Re: Solenoid Controls_Final_Wiring check-out 

Hello Ruben, 

The Instrumentation TEST PLC- EPICS PLAN_v6_June 13th 2017 spreadsheet was update with the new instrumentation that needs to be tested. You can find it on the link : M:\hallb_eng\CLAS12\Magnets\Solenoid\Controls and Wiring. 

The test for the instrumentation, that was required by Scot for the PTS (Temperature Sensors), Cryocon units and solenoid vales will be complete by the end of this week, I will let you know once I complete the tasks. 

Best Regards, 
Pablo 

From: "Ruben Fair" <rfair at jlab.org> 
To: "Pablo Campero Rojas" <campero at jlab.org> 
Sent: Tuesday, June 13, 2017 7:51:22 AM 
Subject: Fwd: Solenoid Controls_Final_Wiring check-out 

Pablo, 

Please confirm that these additional tests have been added to your I and C test spreadsheet. 

When are you planning to complete the tests? 

If you need help from Nick, Scot or myself let us know. 

Regards 

Ruben 


From: "Ruben Fair" <rfair at jlab.org> 
To: "Scot Spiegel" <spiegel at jlab.org> 
Cc: "Probir Ghoshal" <ghoshal at jlab.org>, "Renuka Rajput-Ghoshal" <renuka at jlab.org>, "Pablo Campero Rojas" <campero at jlab.org>, "Nicholas Sandoval" <sandoval at jlab.org> 
Sent: Friday, June 9, 2017 4:12:09 PM 
Subject: Re: Solenoid Controls_Final_Wiring check-out 

Thank you Scot. 

Pablo, 

Can you add these tests to your overall I and C test spreadsheet and then plan to get them completed within the next 2 weeks? 

Regards 

Ruben 



From: "Scot Spiegel" <spiegel at jlab.org> 
To: "Ruben Fair" <rfair at jlab.org> 
Cc: "Probir Ghoshal" <ghoshal at jlab.org>, "Renuka Rajput-Ghoshal" <renuka at jlab.org>, "Pablo Campero Rojas" <campero at jlab.org>, "Nicholas Sandoval" <sandoval at jlab.org> 
Sent: Friday, June 9, 2017 3:13:30 PM 
Subject: Solenoid Controls_Final_Wiring check-out 

All, 

Mark Lester and I have completed the last few cable terminations in the Solenoid Racks. 

The remaining devices will need to verified before the Solenoid arrives. 

Test requirement for - Valve solenoids (vaporizer) 24VDC (X) 3, PT's (X) 2 & CRYOCON #2, Channel D for Vaporizer temperature. 

Best regards, 

Scot Spiegel 
Sr. Technologist / Designer 
Physics / Hall D 
JSA Jefferson Lab 
12000 Jefferson Ave. 
Bldg. 55 Rm. 1546-24 
Newport News, VA. 23606 
Ph. 757-536-5900 
Wk. 757-269-5900 

-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_magnets/attachments/20170616/219072a6/attachment.html>


More information about the Dsg-hallb_magnets mailing list