[Dsg-hallb_magnets] Hall B magnet interlock	and	instrumentation	checkouts complete
    Brian Eng 
    beng at jlab.org
       
    Wed Jul 15 11:45:33 EDT 2020
    
    
  
Looking at the data over the past few years shows some swings to >2. Should those be considered abnormal and mark as fail? If so I think 1-2 would be a safe choice since there are some spikes >1.75 and <1.25 if we used the tighter range, but none since the beginning of the year in which case 1.25-1.75 would be fine.
When the numbers were initially chosen (~07/2018) that sensor was reading ~1.3 (probably during one of the low swings).
I think it depends on the sensor as well as how it fails, basically could either be 0, max, or 555. Which this script would mark as a fail in addition to whatever EPICS alarms there are.
________________________________
From: David Kashy <kashy at jlab.org>
Sent: Wednesday, July 15, 2020 11:20 AM
To: Brian Eng <beng at jlab.org>; Tyler Lemon <tlemon at jlab.org>; Ruben Fair <rfair at jlab.org>; dsg-hallb magnets <dsg-hallb_magnets at jlab.org>
Subject: Re: [Dsg-hallb_magnets] Hall B magnet interlock and instrumentation checkouts complete
Brain,
The set value has remained constant at 1.5 for many years. I would advise to change the range to 1.05 to 2.0. But it could be tighter to help check for operation in a "normal range" then I'd say 1.25 to 1.75.
Is there a known value that is read back when the the transducer or other component in the read-back chain fails?
Thanks
Dave
________________________________
From: Brian Eng <beng at jlab.org>
Sent: Wednesday, July 15, 2020 10:42 AM
To: David Kashy <kashy at jlab.org>; Tyler Lemon <tlemon at jlab.org>; Ruben Fair <rfair at jlab.org>; dsg-hallb magnets <dsg-hallb_magnets at jlab.org>
Subject: Re: [Dsg-hallb_magnets] Hall B magnet interlock and instrumentation checkouts complete
Hi Dave,
This is separate/independent of the alarm values.
The script is used to speed up the instrumentation checkout and simply checks if a sensor is within a range and marks it as pass or fail. Initially every sensor was manually checked on the CSS screen and the value written on paper, but this was extremely time consuming. That was replaced with taking screenshots of the various screens with sensors, but then it was hard to actually see if the sensors were good or bad. This is the current implementation of the instrumentation checkout.
The values used for the range were provided by you years ago and might be different now. I didn't want to change the script to make something a pass if it's actually iffy, also I'm not sure if we bump up the upper limit should the lower follow or do we just increase the upper? For example, since the range for PT8554 is 1-1.5 now we could either do 1-2 or 1.25-1.75 for the limits.
Thanks
Brian
________________________________
From: David Kashy <kashy at jlab.org>
Sent: Wednesday, July 15, 2020 10:33 AM
To: Brian Eng <beng at jlab.org>; Tyler Lemon <tlemon at jlab.org>; Ruben Fair <rfair at jlab.org>; dsg-hallb magnets <dsg-hallb_magnets at jlab.org>
Subject: Re: [Dsg-hallb_magnets] Hall B magnet interlock and instrumentation checkouts complete
Hi Brian,
What type of limits are you talking about. Alarm limits or something else?
Thanks
Dave
________________________________
From: Brian Eng <beng at jlab.org>
Sent: Wednesday, July 15, 2020 10:31 AM
To: David Kashy <kashy at jlab.org>; Tyler Lemon <tlemon at jlab.org>; Ruben Fair <rfair at jlab.org>; dsg-hallb magnets <dsg-hallb_magnets at jlab.org>
Subject: Re: [Dsg-hallb_magnets] Hall B magnet interlock and instrumentation checkouts complete
Should I change the limits for that PT from something other than 1-1.5 then?
________________________________
From: Dsg-hallb_magnets <dsg-hallb_magnets-bounces at jlab.org> on behalf of David Kashy <kashy at jlab.org>
Sent: Wednesday, July 15, 2020 9:53 AM
To: Tyler Lemon <tlemon at jlab.org>; Ruben Fair <rfair at jlab.org>; dsg-hallb magnets <dsg-hallb_magnets at jlab.org>
Subject: Re: [Dsg-hallb_magnets] Hall B magnet interlock and instrumentation checkouts complete
Hi Tyler,
1.5 is the setpoint for control so it will read that most of the time.
Dave
________________________________
From: Tyler Lemon <tlemon at jlab.org>
Sent: Wednesday, July 15, 2020 9:37 AM
To: Ruben Fair <rfair at jlab.org>; David Kashy <kashy at jlab.org>; dsg-hallb magnets <dsg-hallb_magnets at jlab.org>
Cc: Probir Ghoshal <ghoshal at jlab.org>
Subject: Re: [Dsg-hallb_magnets] Hall B magnet interlock and instrumentation checkouts complete
Ruben, Dave,
The "failure" for PT8554 is that the script expects the sensor's reading to be between 1 atma and 1.5 atma. Since the value the script saw for PT8554 was 1.5 atma, it put the status as "FAIL".
Tyler
________________________________
From: Ruben Fair <rfair at jlab.org>
Sent: Wednesday, July 15, 2020 9:27 AM
To: Tyler Lemon <tlemon at jlab.org>
Cc: Probir Ghoshal <ghoshal at jlab.org>
Subject: Fw: [Dsg-hallb_magnets] Hall B magnet interlock and instrumentation checkouts complete
Hi Tyler,
Could you comment on Dave's question below?
Cheers
Ruben
________________________________
From: David Kashy <kashy at jlab.org>
Sent: Wednesday, July 15, 2020 8:46 AM
To: Ruben Fair <rfair at jlab.org>; Probir Ghoshal <ghoshal at jlab.org>; Denny Insley <dinsley at jlab.org>
Cc: Nicholas Sandoval <sandoval at jlab.org>
Subject: Re: [Dsg-hallb_magnets] Hall B magnet interlock and instrumentation checkouts complete
Hi All,
It is correct that TD8522TR has been very unreliable in the past, especially when below 10K. During warm up and cooldown it seems to work better. It is also true that it is not used for control so running without it is not a big deal.
As far as PT8554, that is another story. This sensor is the input parameter for the LN2 reservoir back pressure control. It seems to be working fine at least now when I looked.
What was the indication that caused it to be "failed" during the checkout?
Thanks
Dave
________________________________
From: Ruben Fair <rfair at jlab.org>
Sent: Tuesday, July 14, 2020 3:27 PM
To: David Kashy <kashy at jlab.org>; Probir Ghoshal <ghoshal at jlab.org>; Denny Insley <dinsley at jlab.org>
Cc: Nicholas Sandoval <sandoval at jlab.org>
Subject: Fw: [Dsg-hallb_magnets] Hall B magnet interlock and instrumentation checkouts complete
Dave,
See elog below from the magnet pre-check today - I believe neither the temp or press sensor is being used to control anything?
https://logbooks.jlab.org/entry/3814661
[cid:a085db29-3ee8-43e5-9b24-e8fabd360697]
[cid:9d782124-1718-4ea5-b08f-00251ddc11b5]
Probir,
I think we did at one time try swapping leads to activate the spare TD8522TR but saw the same behavior?
Ruben
________________________________
From: Dsg-hallb_magnets <dsg-hallb_magnets-bounces at jlab.org> on behalf of Tyler Lemon <tlemon at jlab.org>
Sent: Tuesday, July 14, 2020 2:59 PM
To: dsg-hallb magnets <dsg-hallb_magnets at jlab.org>
Subject: [Dsg-hallb_magnets] Hall B magnet interlock and instrumentation checkouts complete
Hello
Pablo, Probir, and I have completed the pre-power-up interlock and instrumentation checkouts for the Hall B Solenoid and Torus. The checkout results have been posted to the logbooks at the links below.
Torus interlock - https://logbooks.jlab.org/entry/3814708
Torus instrumentation - https://logbooks.jlab.org/entry/3814474 , https://logbooks.jlab.org/entry/3814475
Solenoid interlock - https://logbooks.jlab.org/entry/3814709
Solenoid instrumentation - https://logbooks.jlab.org/entry/3814661
Tyler
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_magnets/attachments/20200715/616c7290/attachment-0001.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image.png
Type: image/png
Size: 6442 bytes
Desc: image.png
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_magnets/attachments/20200715/616c7290/attachment-0003.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image.png
Type: image/png
Size: 4359 bytes
Desc: image.png
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_magnets/attachments/20200715/616c7290/attachment-0004.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: PT8554.png
Type: image/png
Size: 18301 bytes
Desc: PT8554.png
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_magnets/attachments/20200715/616c7290/attachment-0005.png>
    
    
More information about the Dsg-hallb_magnets
mailing list