[Dsg-mvt] Request new manometer for Eel premix gas system + interlock

Franck Sabatie sabatie at jlab.org
Mon Jun 12 13:00:51 EDT 2017


Yeah most likely the pressure is too low for that gauge. And we brought the
sniffer back... (it would be useful to have one btw, especially with the
installation of the hall b gas system)

Franck

Le lun. 12 juin 2017 à 18:40, Brian Eng <beng at jlab.org> a écrit :

> Hi Franck,
>
> It could be nearly empty, but it is also just as likely to be that the
> pressure is to low for the gauge.
>
> Generally those are spec'd to a few percent but only in the range of
> 25-75% of the scale, since we already start way below that there's no way
> to really tell exactly what the pressure in the bottle is. Not to mention
> the first tick mark is 200 psi, which isn't much below what the bottle is
> filled to.
>
> Hopefully this will be improved when the 600 psi gauges that Matt approved
> are installed, which is currently in procurement with a buyer assigned.
>
> For the time being you guys can monitor the low pressure side for changes,
> when it can't maintain 15psi is when the bottle is almost empty.
>
> Is your Argon sniffer still at JLAB or did that go back to Saclay?
>
> > On Jun 12, 2017, at 11:48 AM, Franck Sabatie <sabatie at jlab.org> wrote:
> >
> > Dear Marc,
> >
> > It looks nearly empty (or am I reading this wrong?), which is odd : we
> only use 7l/h, a new bottle should last 5-7 days. Maybe there is a leak
> somewhere before or gas rack (we check input vs output flow).
> >
> > Franck
> >
> > Le lun. 12 juin 2017 à 17:45, Marc McMullen <mcmullen at jlab.org> a écrit
> :
> > Hi All,
> >
> > Attached is an updated photo of the MVT regulator from this morning
> (about 10:40am). This bottle was connected around 2pm on Friday.
> >
> > Thanks,
> >
> > Marc
>