[Dsg-mvt] MVT gas mixing electronics/network equipment
Brent Morris
bmorris at jlab.org
Thu Oct 12 14:06:25 EDT 2017
The network cable should be a CAT6 Ethernet cable. If it will be running next to any strong electrical interference, then it should be a shielded CAT6 cable. If any part of the run will expose it to the sun, the outer jacket will need to be outdoor-rated (UV-protected). Don't get a gel-filled cable - since it won't be direct-buried or submerged in water, you shouldn't need it and it is a pain to terminate.
Do you know where the cable will terminate within Hall B? I'd like to make sure we have a switch close to that location that can provide the networking you need.
-Brent
----- Original Message -----
> From: "Marc McMullen" <mcmullen at jlab.org>
> To: "Matthew Marchlik" <marchlik at jlab.org>, "Bob Miller" <rmiller at jlab.org>
> Cc: "Patrizia Rossi" <rossi at jlab.org>, dsg-mvt at jlab.org, "Brent Morris" <bmorris at jlab.org>
> Sent: Wednesday, October 11, 2017 3:31:49 PM
> Subject: Re: [Dsg-mvt] MVT gas mixing electronics/network equipment
> Hi Matt,
>
> I can procure all the cables, but you will need to find manpower for
> installing them (run from the gas pad to the respective locations in
> Hall B). Once they are installed, DSG can schedule in termination and
> testing of the cables.
>
> I will cc Brent Morris on this email so that he can give a
> recommendation for the network cable and connectors for all weather use.
>
> Once the cables and instrumentation (MFCs and transducers) are
> installed, we will schedule the upgrade to the software to run the
> instrumentation using the Hall B gas controls software. Then we will
> begin testing, testing may take up to two weeks.
>
> Thanks
>
> Marc
>
>
> On 10/11/2017 3:03 PM, Matthew Marchlik wrote:
>> Hi Marc,
>> Do you need anything from me or hall B to progress?
>> Matt
>>
>> ----- Original Message -----
>> From: "Marc McMullen" <mcmullen at jlab.org>
>> To: "Matthew Marchlik" <marchlik at jlab.org>
>> Sent: Wednesday, October 11, 2017 11:51:45 AM
>> Subject: Re: [Dsg-mvt] MVT gas mixing electronics/network equipment
>>
>> Hi Matt,
>>
>> I have ordered 3x 1000' spools of cable for the MFC power and Pressure
>> Transducer. It is 3 conductor so it will be fine for both types. I will
>> let you know when it arrives. The cable can not be pre-terminated due to
>> the run through the penetration.
>>
>> We need to have an "in the hall" location for the network cables to
>> terminate and procure network cable for the MFCs.
>>
>> Marc
>>
>>
>> On 10/10/2017 9:57 AM, Matthew Marchlik wrote:
>>> Hi Marc,
>>>
>>> Please fill out/add tasks/comment as you see fit in this table to avid confusion
>>> on who is handling what.
>>>
>>> David and Todd are making good progress on populating the MVT mixing rack and
>>> will get a good portion of the hall tubing routing done this week. I need to
>>> focus on the controls/cabling portion of things so those things are not holding
>>> us up at the end.
>>>
>>> Thanks,
>>> Matt
>>>
>>> ----- Original Message -----
>>> From: "Marc McMullen" <mcmullen at jlab.org>
>>> To: "Matthew Marchlik" <marchlik at jlab.org>
>>> Cc: "Bob Miller" <rmiller at jlab.org>, dsg-mvt at jlab.org, dsg-ft at jlab.org
>>> Sent: Monday, September 25, 2017 1:12:50 PM
>>> Subject: Re: [Dsg-mvt] MVT gas mixing electronics/network equipment
>>>
>>> Matt,
>>>
>>> Later today I will go to the hall and come up with a sketch of the run.
>>> I spoke with Cyril Wiggins last week, he said he could calculate the
>>> lengths. Once I have the lengths, we can start procuring cable. We will
>>> need an account code.
>>>
>>> Marc
>>>
>>>
>>> On 9/25/2017 1:07 PM, Matthew Marchlik wrote:
>>>> Hi Marc,
>>>>
>>>> Is someone already looking into the cable lengths?
>>>>
>>>> Matt
>>>>
>>>> ----- Original Message -----
>>>> From: "Marc McMullen" <mcmullen at jlab.org>
>>>> To: "Matthew Marchlik" <marchlik at jlab.org>
>>>> Cc: "Bob Miller" <rmiller at jlab.org>, dsg-mvt at jlab.org, dsg-ft at jlab.org
>>>> Sent: Monday, September 25, 2017 12:59:16 PM
>>>> Subject: Re: [Dsg-mvt] MVT gas mixing electronics/network equipment
>>>>
>>>> Hi Matt,
>>>>
>>>> The chassis will go through continuity and power testing this week. At
>>>> that point it can be installed in the SFL3 Rack.
>>>>
>>>> As far as the cables, I am still have not received information on
>>>> lengths for the cable runs. So that the cable can be procured.
>>>>
>>>> Can you provide some scheduling information for the MVT installation.
>>>> The DSG manager will need this information for our own scheduling.
>>>>
>>>> Thanks,
>>>>
>>>> Marc
>>>>
>>>>
>>>> On 9/25/2017 11:35 AM, Matthew Marchlik wrote:
>>>>> Hi Marc,
>>>>>
>>>>> I'm looking for an update on some items:
>>>>>
>>>>> Has the interface chassis been relocated to SFL3?
>>>>>
>>>>> Have the five MFC power cables been run from SFL3 to the pad?
>>>>>
>>>>> Have the PT power and signal cables been run?
>>>>>
>>>>> Thanks,
>>>>> Matt
>>>>>
>>>>>
>>>>> ----- Original Message -----
>>>>> From: "Marc McMullen" <mcmullen at jlab.org>
>>>>> To: "Bob Miller" <rmiller at jlab.org>, "Matthew Marchlik" <marchlik at jlab.org>
>>>>> Cc: dsg-mvt at jlab.org, dsg-ft at jlab.org
>>>>> Sent: Friday, September 15, 2017 12:25:14 PM
>>>>> Subject: Re: [Dsg-mvt] MVT gas mixing electronics/network equipment
>>>>>
>>>>> Hi Matt
>>>>>
>>>>> Relocation of the MVT mixing system controls to accommodate the
>>>>> installation of the MVT mixing system on the Hydrogen Target gas pad.
>>>>>
>>>>> *MFC Power:*
>>>>>
>>>>> The mass flow controllers will be located outside. To ensure power
>>>>> during outages we will need to locate the controls interface chassis in
>>>>> Hall B on SFL3. Five MFC power cables should be run from the SFL3_south
>>>>> through the penetration to the gas pad.
>>>>>
>>>>> *MFC Communication:*
>>>>>
>>>>> Per our conversation with Brent Morris, Hall B will need to run network
>>>>> cable from an available UPS backed up network switch on the appropriate
>>>>> subnet to the gas pad through the penetration.
>>>>>
>>>>> *Signals:*
>>>>>
>>>>> Pressure: The MVT has two pressure transducers which will be located on
>>>>> the HT gas pad. Power and signal cable should follow the same route as
>>>>> the MFC power to SFL3_south.
>>>>>
>>>>> Since the MFCs are temperature sensitive we should add redundant
>>>>> Temperature monitoring to the new MFC housing. These signals can be
>>>>> monitored by the Hall B Gas Controls software and provide that data to
>>>>> EPICS.
>>>>>
>>>>> Type
>>>>>
>>>>>
>>>>>
>>>>> # of cables
>>>>>
>>>>>
>>>>>
>>>>> Start Location
>>>>>
>>>>>
>>>>>
>>>>> Comment
>>>>>
>>>>> MFC Power
>>>>>
>>>>>
>>>>>
>>>>> 5
>>>>>
>>>>>
>>>>>
>>>>> SFL3_S
>>>>>
>>>>>
>>>>>
>>>>> Expected (I) 0.167A. Shielded 2cond. 22 awg.
>>>>>
>>>>> MFC Communication
>>>>>
>>>>>
>>>>>
>>>>> 5
>>>>>
>>>>>
>>>>>
>>>>> Hall B
>>>>>
>>>>>
>>>>>
>>>>> Cat. 5
>>>>>
>>>>> Pressure Transducers
>>>>>
>>>>>
>>>>>
>>>>> 2
>>>>>
>>>>>
>>>>>
>>>>> SFL3_S
>>>>>
>>>>>
>>>>>
>>>>> Expected (I) 0.140A. shielded 3cond. 22 awg.
>>>>>
>>>>> MFC Temp monitor
>>>>>
>>>>>
>>>>>
>>>>> 2
>>>>>
>>>>>
>>>>>
>>>>> SFL3_S
>>>>>
>>>>>
>>>>>
>>>>> TBD
>>>>>
>>>>> Hall B should provide the cable length for these runs so that DSG can
>>>>> procure the necessary cable. The previous length expectation was much
>>>>> shorter.
>>>>>
>>>>> Regards,
>>>>>
>>>>> Marc
>>>>>
>>>>>
>>>>> On 9/14/2017 3:59 PM, Marc McMullen wrote:
>>>>>> Hi Matt,
>>>>>>
>>>>>> I'm scheduled to meet with DSG management in the morning to discuss a
>>>>>> strategy for the relocation of the controls equipment. We need to make
>>>>>> sure we cover all bases with power, network, UPS, and temperature
>>>>>> monitoring of the MFCs. We will get back to you tomorrow on the matter.
>>>>>>
>>>>>> Regards,
>>>>>>
>>>>>> Marc
>>>>>>
>>>>>>
>>>>>>
>>>>>> On 9/14/2017 11:34 AM, Matthew Marchlik wrote:
>>>>>>> Hi Marc,
>>>>>>>
>>>>>>> I met with Walt Akers, Tim Minga and Paul Powers this morning to
>>>>>>> discuss installation of the system into 96B. The consensus is that
>>>>>>> there is not way to modify the shed for flammable gas service without
>>>>>>> spending significant time and money. Installing into 96C, hall C's
>>>>>>> gas shed, was also discussed but would not work out.
>>>>>>>
>>>>>>> So, the mixing system and gas supply bottles will be installed on the
>>>>>>> gas pad. Could you please reply to this email and list the
>>>>>>> items/steps necessary for installation from DSG's standpoint we
>>>>>>> discussed yesterday?
>>>>>>>
>>>>>>> Hall B will create a small insulated foam box to house the MFCs with
>>>>>>> a segment of directly controlled heat tape within to keep the
>>>>>>> temperature above 50F. It will have holes for the cables. Is there
>>>>>>> any need to regularly access the MFCs after commissioning?
>>>>>>>
>>>>>>> The C4H10 line from the bottle to the MFCs will need to be wrapped
>>>>>>> with insulation and heat tape set to 90F, same as the bottle blanket.
>>>>>>> (George, can this be powered/controlled in series with the heater
>>>>>>> blanket?)
>>>>>>>
>>>>>>> Hall B and I will update all of the PS documentation, drawings,
>>>>>>> models and P&IDs.
>>>>>>>
>>>>>>> Thanks,
>>>>>>> Matt
>>>>>>>
>>>>>>>
>>>>>>> ----- Original Message -----
>>>>>>> From: "Marc McMullen" <mcmullen at jlab.org>
>>>>>>> To: "Matthew Marchlik" <marchlik at jlab.org>, "Bob Miller"
>>>>>>> <rmiller at jlab.org>, dsg-mvt at jlab.org
>>>>>>> Sent: Wednesday, September 6, 2017 3:53:29 PM
>>>>>>> Subject: MVT gas mixing electronics/network equipment
>>>>>>>
>>>>>>> Hi Matt and Bob,
>>>>>>>
>>>>>>> Is there a firm plan on the location of the MVT mixing instrumentation,
>>>>>>> electronics, and network equipment. If the equipment cannot be located
>>>>>>> in the Hall, I understand that the gas pad may be the next logical
>>>>>>> choice. The instrumentation which will have to be at the gas panel are
>>>>>>> the MFCs, which are temperature sensitive. The MFCs communicate via the
>>>>>>> network, so under normal circumstances we would like to have a switch at
>>>>>>> the location. We would also like to put the power for the MFCs (a
>>>>>>> chassis containing a power supply) in that location if possible. To
>>>>>>> house all of these components, may require large enclosure, or rack
>>>>>>> which can be enclosed and climate controlled to keep all components
>>>>>>> running and in good shape for the long run. At any rate, please let us
>>>>>>> (DSG) know as soon as a decision is made, so that we can plan
>>>>>>> accordingly.
>>>>>>>
>>>>>>>
>>>>>>> Thanks,
>>>>>>>
>
> --
> Marc McMullen
> Detector Support Group
> 757-269-7738
More information about the Dsg-mvt
mailing list