[Dsg-hallb_prad_ii] Labview program data logging
    Xiangdong Wei 
    xwei at jlab.org
       
    Fri Jun 13 17:03:52 EDT 2025
    
    
  
Hi Pablo,
Thanks for tour message.
I have started the cooldown process when seeing your email. I'll test the program on Monday when the changing the run conditions.
Best,
Xiangdong
________________________________
From: Pablo Campero Rojas <campero at jlab.org>
Sent: Friday, June 13, 2025 3:04 PM
To: Xiangdong Wei <xwei at jlab.org>; Tsuneo Kageya <kageya at jlab.org>; Phillip Dobrenz <dobrenz at jlab.org>; Andrew Schick <aschick at umass.edu>; dsg-hallb_prad_ii at jlab.org <dsg-hallb_prad_ii at jlab.org>
Subject: Re: [Dsg-hallb_prad_ii] Labview program data logging
Hello Xiangdong,
The new version 2 of the local data logger LabVIEW code was implemented in the main PRad LabVIEW program and it is available in the test stand computer at: C:\Users\prad-ii\Desktop
The name of the folder with the LabVIEW program is: PRadII-June-13-2025
You can use and test this newer version of the code to archive the data and limit the size of the automatically created Excel files. The size of each Excel file can be set by the expert user in the LabVIEW front panel by entering the numbers of minutes to archive data in each Excel file.
As we discussed, some of the LabVIEW indicators do not lid since there is no read back from the NI-6001 DAQ digital input channels, this can be checked next week when the testing is not running, just let me know.
Let me know if you have any questions.
Regards,
Pablo
________________________________
From: Dsg-hallb_prad_ii <dsg-hallb_prad_ii-bounces at jlab.org> on behalf of Pablo Campero Rojas <campero at jlab.org>
Sent: Wednesday, June 11, 2025 12:44 PM
To: Xiangdong Wei <xwei at jlab.org>; Tsuneo Kageya <kageya at jlab.org>; Phillip Dobrenz <dobrenz at jlab.org>; Andrew Schick <aschick at umass.edu>; dsg-hallb_prad_ii at jlab.org <dsg-hallb_prad_ii at jlab.org>
Subject: Re: [Dsg-hallb_prad_ii] Labview program data logging
Hello Xiangdong,
Yes, I can check these items in the upcoming days.
Regards,
Pablo
________________________________
From: Xiangdong Wei <xwei at jlab.org>
Sent: Wednesday, June 11, 2025 12:11 PM
To: Pablo Campero Rojas <campero at jlab.org>; Tsuneo Kageya <kageya at jlab.org>; Phillip Dobrenz <dobrenz at jlab.org>; Andrew Schick <aschick at umass.edu>
Subject: Re: [Dsg-hallb_prad_ii] Labview program data logging
Hi Pablo,
There are a few accumulated issues with the PRad-II operation, listed below, which need to be addressed soon. Part of them is LabVIEW related, I think.
  1.
Data recording stops after 4 hours (that's planned, I think), but no new file started.
  2.
The 3 am IT scan will restart the recording, but non-numbers were recorded due to the loss of USB connection.
  3.
At 6:30pm every day, the data recording will stop. Again, IT scan will start recording junks.
  4.
Pressure sensors and switches were not all calibrated.
  5.
Some LabVIEW buttons are not lid when clicking them.
  6.
Some turbo pumps loss speed at moderate flow (for target expert to check).
  7.
...
Would you please check the items 1, 3 and 5? The target is at room temperature for the next few days.
Phillip will contact the IT for adding the PC to the experimental network.
I'll talk to James Brock and Chris Keith this afternoon for item 4 and 6.
Thanks,
Xiangdong
________________________________
From: Dsg-hallb_prad_ii <dsg-hallb_prad_ii-bounces at jlab.org> on behalf of Xiangdong Wei <xwei at jlab.org>
Sent: Friday, May 30, 2025 7:38 PM
To: Pablo Campero Rojas <campero at jlab.org>; Tsuneo Kageya <kageya at jlab.org>; dsg-hallb_prad_ii at jlab.org <dsg-hallb_prad_ii at jlab.org>
Subject: Re: [Dsg-hallb_prad_ii] Labview program data logging
Gentlemen,
I realized that the interlock PS1 was a replacement piece which was not set to correct value yet, which kept tripping when helium was flowed into target cell.
I then changed the tripping threshold to 1 torr and tested the flow for up to 600 SCCM and the interlock held pretty well. I cleaned up the target vessel with all five turbo pumps for about 24 hours and the started the refrigerator with 110 SCCM flow, the minimum flow rate we can settle to, around 5 PM today.
If the target is still cold on Monday, we will start increasing the flow rate and then finish our original planned tests.
Another thing I realized that during the hidden test in target area, the fifth turbo pump was never used, and the fifth turbo pump only saw hydrogen gas in the hall. The complete system we have right now at the ESB is more than either the run in the hall or the test at the target area. If we really want to repeat the original test, we will need to remove the fifth turbo pump from the current setup. We’ll see if necessary after next week’s flow test.
The relevant logbook pages and some screenshots are attached.
Thanks,
Xiangdong
Get Outlook for iOS<https://aka.ms/o0ukef>
________________________________
From: Xiangdong Wei <xwei at jlab.org>
Sent: Wednesday, May 28, 2025 12:08:22 PM
To: Pablo Campero Rojas <campero at jlab.org>; Tsuneo Kageya <kageya at jlab.org>; dsg-hallb_prad_ii at jlab.org <dsg-hallb_prad_ii at jlab.org>
Subject: Re: Labview program data logging
Hi Pablo,
Got it.
Thanks,
Xiangdong
________________________________
From: Pablo Campero Rojas <campero at jlab.org>
Sent: Wednesday, May 28, 2025 11:52 AM
To: Xiangdong Wei <xwei at jlab.org>; Tsuneo Kageya <kageya at jlab.org>; dsg-hallb_prad_ii at jlab.org <dsg-hallb_prad_ii at jlab.org>
Subject: Re: Labview program data logging
Hello Xiangdong,
The data acquisition and control response is not affected by the data logging interval set by the user.
I can take a look the code this afternoon after my DSG weekly meeting.
I'm also planing on testing the second version of the local data logger code which will limit the Excel files size to store only 6 hrs of data as we agreed.
Regards,
Pablo
________________________________
From: Xiangdong Wei <xwei at jlab.org>
Sent: Wednesday, May 28, 2025 10:55 AM
To: Pablo Campero Rojas <campero at jlab.org>; Tsuneo Kageya <kageya at jlab.org>; dsg-hallb_prad_ii at jlab.org <dsg-hallb_prad_ii at jlab.org>
Subject: Re: Labview program data logging
Hi Pablo,
There are a few minor issues with the Labview control program which would need to be addressed:
  1.
The logfile (excel) on 20250523 shown the recording stopped at 6:30pm, and resumed automatically at 3am next day, after IT scanned the PC. That was about ~3,000 lines in the logfile at 6:30pm.
  2.
The data logging interval seemed equals to the acquisition interval, which would reduce the control response and potentially damage the equipment (if the interval was set too long).
Please take a look.
Thanks,
Xiangdong
________________________________
From: Pablo Campero Rojas <campero at jlab.org>
Sent: Friday, May 23, 2025 12:29 PM
To: Tsuneo Kageya <kageya at jlab.org>; dsg-hallb_prad_ii at jlab.org <dsg-hallb_prad_ii at jlab.org>
Cc: Xiangdong Wei <xwei at jlab.org>
Subject: Re: Labview program data logging
Hello Tsuneo,
Here is a couple of items that can be checked:
  1.
Make sure that interlock chassis has no faults
  2.
Ensure that button with label “Vac Sw Set/Reset” in the interlock chassis is enabled
  3.
To control EV1 in local mode, Control Mode button on the Valve Control panel must be disabled (no light on button)
     *
Press the "Open" button on the Valve Control panel
     *
Check the "Open" button light indicator, it should stay latched and display green color
  4.
To control EV1 remotely, make sure the Control Mode button on the Valve control panel is enabled (Orange color = Remote control On) and the LED indicator labeled "Remote Signal Enabled" is On
     *
In LabVIEW user interface make sure the "AIO" communication status indicator is green, if not, reset communication with DAQ device.
     *
Control EV1 valve aperture using controls in LabVIEW user interface
  5.
After checking these few steps, the local or remote controls should work, if not, let me know and we can meet when you're done running your test with EV1 closed.
Regards,
Pablo
________________________________
From: Tsuneo Kageya <kageya at jlab.org>
Sent: Friday, May 23, 2025 11:48 AM
To: Pablo Campero Rojas <campero at jlab.org>; dsg-hallb_prad_ii at jlab.org <dsg-hallb_prad_ii at jlab.org>
Cc: Xiangdong Wei <xwei at jlab.org>; Tsuneo Kageya <kageya at jlab.org>
Subject: Re: Labview program data logging
Hola Pablo,
 yesterday, we could not open EV1 remotely nor locally.    I think the same thing happened when you were there and fixed it.    If you know how to fix it,
please let us know.      Now we are running without He gas flow and would not
like to open EV1, but we can open it the next week.
      Saludos, Tsuneo Kageya.
________________________________
From: Pablo Campero Rojas <campero at jlab.org>
Sent: Wednesday, May 21, 2025 9:46 AM
To: Xiangdong Wei <xwei at jlab.org>; dsg-hallb_prad_ii at jlab.org <dsg-hallb_prad_ii at jlab.org>
Cc: Tsuneo Kageya <kageya at jlab.org>
Subject: Re: Labview program data logging
Hello Xiandong,
The LabVIEW code is fine, this is how it normally works, you can't open the excel file to which LABVIEW is writing the data while it is running. There is no way of having two sources writing data to the same file simultaneously, that is why as soon as you open that excel file it displays an error.
The solution to look at the data while data logging program is running as follow:
1.Make a copy of the excel file with the recorded data, location for the excel file is shown in LabVIEW File Path indicator.
2. Open this copy, it will contain the data until the second when the copy was made
3. If desired, after checking the data you can deleted this copy, so we do not take unnecessary space on the local C drive
You can repeat these steps to look at the data while data keeps recording as many times as you want. I hope the explanation is clear, if not, let me know and I can clarify your questions.
Also, when I came this morning to the ESB, I noticed that the SPFDOBRENZ computer was sleeping and suspended.
I'm still awaiting a response from the computer center to resolve this issue.
Regards,
Pablo
________________________________
From: Xiangdong Wei <xwei at jlab.org>
Sent: Wednesday, May 21, 2025 1:16 AM
To: Pablo Campero Rojas <campero at jlab.org>
Cc: Phillip Dobrenz <dobrenz at jlab.org>; Tsuneo Kageya <kageya at jlab.org>; Andrew Schick <aschick at umass.edu>
Subject: Re: Labview program data logging
Gentlemen,
The PRad2 target trip a few more times this afternoon, and the tripping cycle got shorter and shorter. The last time was only three minutes. In the same time, PT-2 went up to 12 psi and the labview dropped flow rate to 0.
All symptoms indicated that the floor line was blocked.
Tsuneo and I decided to warm up the target for another round of the leak checking to determine the cost of blockage.
Pablo, it looks like if we check the excel file without turning off the data logging, the labview program will crash (stuck on saving data). Could you take a look at your convenient time?
Thanks,
Xiangdong
Get Outlook for iOS<https://aka.ms/o0ukef>
________________________________
From: Pablo Campero Rojas <campero at jlab.org>
Sent: Tuesday, May 20, 2025 8:17:58 AM
To: Xiangdong Wei <xwei at jlab.org>
Cc: Phillip Dobrenz <dobrenz at jlab.org>; Tsuneo Kageya <kageya at jlab.org>; Andrew Schick <aschick at umass.edu>
Subject: Re: Labview program data logging
Hello Xiangdong,
I will be there around 10 am to check the issues with the code.
Regards,
Pablo
________________________________
From: Xiangdong Wei <xwei at jlab.org>
Sent: Monday, May 19, 2025 11:59:36 PM
To: Pablo Campero Rojas <campero at jlab.org>
Cc: Phillip Dobrenz <dobrenz at jlab.org>; Tsuneo Kageya <kageya at jlab.org>; Andrew Schick <aschick at umass.edu>
Subject: Labview program data logging
Hi Pablo,
Today we found that the live numbers and plots in PRad-II control program were not updated when the data logging button was enabled. The excel spreadsheet showed all lines were just the repeats of the first line. The Excel file you saved when testing the new feature also shown the repeating data set. The data display went back to alive after logging stopped.
Please take a look at the program and fix this issue.
We are running the PRad-II target this week. The target temperature was set to 15K for overnight. We will be there after 9 AM tomorrow.
Thanks,
Xiangdong
Get Outlook for iOS<https://aka.ms/o0ukef>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_prad_ii/attachments/20250613/b66940f0/attachment-0001.htm>
    
    
More information about the Dsg-hallb_prad_ii
mailing list