[Dsg-rich] corrected - Re: Upgrade to the RICH Valve panel

Marc McMullen mcmullen at jlab.org
Tue Dec 5 14:20:11 EST 2017


Hi Matt,

Can you give a status on the finalization of the Pressure Systems sign 
off? George is working on the individual components (manifold) for the 
upgraded parts. We will need to start the installation by sometime on 
Thursday. The upgrades must be done in time for the Aerogel installation 
next week. Let me know if there is anything else I can provide to 
expedite the process.


Thanks,

Marc


On 12/4/2017 11:29 AM, Marc McMullen wrote:
> Hi Matt,
>
> This is for the EEL. George says the relief valves that come with the 
> dewars are 230 - 240psi.
>
> Marc
>
>
> On 12/4/2017 9:38 AM, Matthew Marchlik wrote:
>> Hi Marc,
>>
>> The original panel P&ID only included a "Hall B N2 supply" note and 
>> didn't include the dewars. Why show them now? Are they the same as 
>> the original dewar, capable of providing 65 psig?
>>
>> Matt
>>
>> ----- Original Message -----
>> From: "Marc McMullen" <mcmullen at jlab.org>
>> To: "Matthew Marchlik" <marchlik at jlab.org>, "dsg-rich" 
>> <dsg-rich at jlab.org>
>> Sent: Thursday, November 30, 2017 12:09:17 PM
>> Subject: Re: corrected - Re: [Dsg-rich] Upgrade to the RICH Valve panel
>>
>> Hi Matt,
>>
>> Here are the latest upgrades to the P&I diagram. None of the filters
>> have been installed yet. The N2 sketch includes plans to run up to four
>> dewars through a manifold (four would only be needed during Christmas
>> Break) . Please feel free to let me know of any changes needed. Thanks
>> for expediting the matter.
>>
>>
>> Regards,
>>
>> Marc
>>
>>
>> On 11/30/2017 8:54 AM, Matthew Marchlik wrote:
>>> Hi Marc,
>>>
>>> What is the status of the RICH alterations?
>>>
>>> As a reminder, alterations should not be energized until I have 
>>> signed things off. I could still use the corrected P&IDs.
>>>
>>> Thanks,
>>> Matt
>>>
>>>
>>> ----- Original Message -----
>>> From: "Marc McMullen" <mcmullen at jlab.org>
>>> To: "Matthew Marchlik" <marchlik at jlab.org>
>>> Cc: "Bob Miller" <rmiller at jlab.org>, dsg-rich at jlab.org
>>> Sent: Monday, November 20, 2017 4:15:33 PM
>>> Subject: Re: corrected - Re: [Dsg-rich] Upgrade to the RICH Valve panel
>>>
>>> Will do.
>>>
>>>
>>> On 11/20/2017 4:13 PM, Matthew Marchlik wrote:
>>>> Marc,
>>>>
>>>> For the compressed air P&ID:
>>>>
>>>> Please eliminate the "TO RICH #1 & #2 ATM" notes.
>>>>
>>>> Add two dashed boxes to show boundary of RICH detectors and add the 
>>>> manifolds.
>>>>
>>>> Add isolation valves where needed.
>>>>
>>>> Move the pressure system boundary to encompass the manifolds.
>>>>
>>>>
>>>> For the N2 P&ID:
>>>> Add the isolation valve.
>>>>
>>>> Matt
>>>>
>>>> ----- Original Message -----
>>>> From: "Marc McMullen" <mcmullen at jlab.org>
>>>> To: "Bob Miller" <rmiller at jlab.org>, "Matthew Marchlik" 
>>>> <marchlik at jlab.org>
>>>> Cc: dsg-rich at jlab.org
>>>> Sent: Thursday, November 16, 2017 2:09:15 PM
>>>> Subject: corrected - Re: [Dsg-rich] Upgrade to the RICH Valve panel
>>>>
>>>> I have updated the p/ns on the diagram. The new parts are in red.
>>>>
>>>> N2 Filter
>>>> <https://atlascopcocompressorstore.com/21-cfm-coalescing-filter-1-micron-38-npt245> 
>>>>
>>>> Air Filter
>>>> <https://atlascopcocompressorstore.com/106-cfm-coalescing-filter-001-micron-1-npt> 
>>>>
>>>> On 11/16/2017 11:38 AM, Marc McMullen wrote:
>>>>> Hi Bob and Matt,
>>>>>
>>>>> I have attached two diagrams with the changes which we need to make
>>>>> today. The N2 change is the most critical, due to the installation of
>>>>> the Aerogel. I will contact you this morning to discuss any issues 
>>>>> you
>>>>> may have with the change.
>>>>>
>>>>> Thanks,
>>>>>
>>>>> Marc
>>>>>
>>>>>
>>>>> On 11/16/2017 9:47 AM, Marc McMullen wrote:
>>>>>> Hi Bob,
>>>>>>
>>>>>> We are preparing to start continuous N2 flow to the RICH. Once the
>>>>>> Aerogel is installed the purge must be continuous to avoid the
>>>>>> absorption of moisture. Before this happens we need to install an 
>>>>>> oil
>>>>>> coalescing filter before the N2 supply valve to the panel. We also
>>>>>> have been requested to install an oil coalescing hepa-filter on the
>>>>>> output of the compressor (we have purchased filters for both
>>>>>> compressors).
>>>>>>
>>>>>> We also have been requested to monitor the moisture content of the
>>>>>> compressed air, to do this we will use the same instrument used on
>>>>>> the DC gas. George is looking into the best installation location on
>>>>>> the valve panel for the moisture sensor.
>>>>>>
>>>>>> Per the Jlab pressure systems program we need to run this through 
>>>>>> you
>>>>>> (as the system owner) and get all the approvals before modifying the
>>>>>> approved system. We should have an updated P&I ready today.
>>>>>>
>>>>>> thanks,
>>>>>>
>>>>>> Easidew Moisture 
>>>>>> <http://www.michell.com/us/products/easidew-online.htm>
>>>>>>
>>>>>> N2 Filter
>>>>>> <https://atlascopcocompressorstore.com/42-cfm-coalescing-filter-001-micron-12-npt?ads_cmpid=775074235&ads_adid=41047268336&ads_matchtype=&ads_network=g&ads_creative=185515089115&utm_term=&ads_targetid=pla-18283950120&utm_campaign=&utm_source=adwords&utm_medium=ppc&ttv=2&gclid=Cj0KCQiAi7XQBRDnARIsANeLIevHsCtFXrb7Bgf9-1GxmhW3gKQnEZATB0nAiLJAIJ1H6kxDYcOYMscaAgSPEALw_wcB> 
>>>>>>
>>>>>>
>>>>>> Air Filter
>>>>>> <https://atlascopcocompressorstore.com/42-cfm-coalescing-filter-001-micron-12-npt?ads_cmpid=775074235&ads_adid=41047268336&ads_matchtype=&ads_network=g&ads_creative=185515089115&utm_term=&ads_targetid=pla-18283950120&utm_campaign=&utm_source=adwords&utm_medium=ppc&ttv=2&gclid=Cj0KCQiAi7XQBRDnARIsANeLIevOxdc2XtpKUZ-aCs_1uH_NjE4QghCQJHMbdMC0gJbH6zvI2Z4jGb0aArY-EALw_wcB> 
>>>>>>
>>>>>>
>>>>>>
>>>>>> <https://atlascopcocompressorstore.com/42-cfm-coalescing-filter-001-micron-12-npt?ads_cmpid=775074235&ads_adid=41047268336&ads_matchtype=&ads_network=g&ads_creative=185515089115&utm_term=&ads_targetid=pla-18283950120&utm_campaign=&utm_source=adwords&utm_medium=ppc&ttv=2&gclid=Cj0KCQiAi7XQBRDnARIsANeLIevOxdc2XtpKUZ-aCs_1uH_NjE4QghCQJHMbdMC0gJbH6zvI2Z4jGb0aArY-EALw_wcB> 
>>>>>>
>>>>>>
>>>>>>
>>>>>> -- 
>>>>>> Marc McMullen
>>>>>> Detector Support Group
>>>>>> 757-269-7738
>>>>>>
>>>>>>
>>>>>> _______________________________________________
>>>>>> Dsg-rich mailing list
>>>>>> Dsg-rich at jlab.org
>>>>>> https://mailman.jlab.org/mailman/listinfo/dsg-rich
>>>>> -- 
>>>>> Marc McMullen
>>>>> Detector Support Group
>>>>> 757-269-7738
>

-- 
Marc McMullen
Detector Support Group
757-269-7738




More information about the Dsg-rich mailing list