[Dsg-rich] humidity sensor calibration
Tyler Lemon
tlemon at jlab.org
Wed Nov 29 09:34:16 EST 2017
Hello Marco,
We would be available to meet this morning at 10:00AM.
So DSG can prepare for the meeting, what issue is there with the readings during the time interval in your logbook entry from 2017-11-25 from 14:00 to 18:00?
Taking the data for that time interval, we wrote a Python script that parses the data every 10 seconds and calculates the maximum difference between humidity readings (1440 data points total). The resulting plot (2017-11-25_1400_to 1800.jpeg) is attached. From the data, the largest difference is 1.22% and the average difference is 0.75% +/- 0.22%. This is within the specs of the sensor.
If the issue is the fluctuations in humidity readings in the middle of the plot, that seems to be a result of whatever programming the drybox uses to reach a specified humidity. It is not from the humidity sensors.
Best regards,
Tyler
From: "Mirazita Marco" <Marco.Mirazita at lnf.infn.it>
To: "dsg-rich" <dsg-rich at jlab.org>
Sent: Tuesday, November 28, 2017 11:15:00 AM
Subject: [Dsg-rich] humidity sensor calibration
Hi all,
I would have a meeting to discuss the calibration of the humidity
sensors at values below 10%.
The data I took few days ago can be found here:
https://logbooks.jlab.org/entry/3496207
When would you be available?
Thank you,
Marco
_______________________________________________
Dsg-rich mailing list
Dsg-rich at jlab.org
https://mailman.jlab.org/mailman/listinfo/dsg-rich
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_rich/attachments/20171129/51f21c23/attachment-0002.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: 2017-11-25_1400_to_1800.jpeg
Type: image/jpeg
Size: 263775 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_rich/attachments/20171129/51f21c23/attachment-0002.jpeg>
More information about the Dsg-rich
mailing list