[Dsg-rich] Assumptions for the N2 purge test
Pablo Campero
campero at jlab.org
Mon Oct 16 13:56:04 EDT 2017
Hello,
Some assumptions for the RICH nitrogen purge test by using Nitrogen cylinder gas as supply:
RICH nitrogen volume: 5000 [l]
Inlet Nitrogen pipe: 3/8 ''
Inside Nitrogen pipe: 1/4 '' (6 pipes per inlet line)
Outlet Nitrogen pipe: 3/4 ''
Area for 3/8 ''-pipe: 0.11 [in^2]
Area for 1/4''-pipe (x 6): 0. 29[in ^2]
Ratio (Inside/Inlet area): 2.64
Max allowed pressure: 15 [psi] (Pressure after the regulator in the N2 gas cylinder)
Inside RICH pressure: 14.69 [psi] (atmospheric pressure)
Calculated flow rate: ~ 250 [LPM]
Time to supply the entire RICH volume with nitrogen: ~ 20 min/line (If two lines are used ~10 [min])
Based on the results, there won't back pressure, and we'll be able to flow the nitrogen at ~ 250 [l/min] without exceeding the established max pressure (15 [psi]).
Best Regards,
Pablo