valery, 
as marc's email yesterday suggested, keeping the roll up door open increases the temp in the hall. 
tyler looked into the matter today and noticed that the hall temp goes up by 5 F. 
(tyler will send a detailed report after the dsg meeting tomorrow) 
so for now the options are: 


    1. ask that the roll up door be kept shut, as much as possible . 
    2. increase air flow by about 5% 
    3. increase threshold by another 3 C. 

in reality we have 3C3+3C2+3C1 +3C0 options! 
amrit 


From: vpk at jlab.org 
To: dsg-rich at jlab.org, vpk at jlab.org 
Sent: Tuesday, August 7, 2018 4:00:04 PM 
Subject: [Dsg-rich] [New Logentry] RICH high temperature interlock came off again. 

RICH high temperature interlock came off again. 


Lognumber 3584622 . Submitted by vpk on Tue, 08/07/2018 - 15:58. Logbooks: 	RICH 

Entry Makers: 	vpk 


We have the same situation today around noon: high temp interlock (N5 and 6) switched RICH off lv and hv. 
I increased (+2C) the high limits at these sensors. It is still acceptable for the front-end electronics. 

_______________________________________________ 
Dsg-rich mailing list 
Dsg-rich at jlab.org 
https://mailman.jlab.org/mailman/listinfo/dsg-rich 
G 
M 
T 
		
		
	
	
	
	
Text-to-speech function is limited to 200 characters 
	
	Options : History : Feedback : Donate 	Close 
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_rich/attachments/20180807/3bc549bb/attachment-0002.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Screen Shot 2018-08-07 at 15.49.30.png
Type: image/png
Size: 366849 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_rich/attachments/20180807/3bc549bb/attachment-0002.png>


More information about the Dsg-rich mailing list