Because of misread of the pressure transducer's (PT) manual, the PT was connected incorrectly to 5 V source when it requires a 24 V power source. 

The wiring of the PT has been corrected and the N2 cRIO is back running in normal operation. The PT was changed to be read by N2 cRIO module 7, channel 4. 

The PT is now measuring the N2 panel supply pressure to be ~ 38 psi. 