[Dsg-rich] [Rich] FPGA temp alarm
Valery Kubarovsky
vpk at jlab.org
Tue Oct 16 09:40:12 EDT 2018
The detector (actually LV in this particular case, because the HV was off at that moment)
was shut down by software interlock that monitors the FPGS temperature.
> From: "Marco Mirazita" <Marco.Mirazita at lnf.infn.it>
> To: "Tyler Lemon" <tlemon at jlab.org>
> Cc: "Valery Kubarovsky" <vpk at jlab.org>, rich at jlab.org, "dsg-rich"
> <dsg-rich at jlab.org>
> Sent: Tuesday, October 16, 2018 9:37:41 AM
> Subject: Re: [Rich] [Dsg-rich] FPGA temp alarm
> But we have a soft interlock shutting down the system if fpga
> temperature goes above 75 deg.
> Il 2018-10-16 14:24 Tyler Lemon ha scritto:
> > Hello Marco,
> > The hardware interlock system does not monitor FPGA temperatures. It
> > monitors the RTDs installed in the EP.
> > The hardware interlock system did not trip off the electronics because
> > the strange temperature readings were only seen on the FPGA
> > temperature readouts.
> > Best regards,
> > Tyler
> > -------------------------
> > FROM: "Mirazita Marco" <Marco.Mirazita at lnf.infn.it>
> > TO: "Valery Kubarovsky" <vpk at jlab.org>
> > CC: rich at jlab.org, "dsg-rich" <dsg-rich at jlab.org>
> > SENT: Tuesday, October 16, 2018 8:16:06 AM
> > SUBJECT: Re: [Rich] [Dsg-rich] FPGA temp alarm
> > Hi Valery,
> > thank you, in fact my main worry was that the interlock didn't shut
> > down
> > the electronics.
> > It would be anyway good to understand why we had this event.
> > I saw from the strip charts that also the LV currents and voltages
> > have
> > frequent random spikes.
> > Marco
> > Il 2018-10-16 13:55 Valery Kubarovsky ha scritto:
> >> Marco,
> >> If you take a more careful look you will find out that the LV was
> > shut
> >> down.
> >> The HV was off. We investigate the reason. It is the first time we
> > had
> >> such an event.
> >> Probably it was connected with the software update that was done
> >> approximately at the same time.
> >> Regards,
> >> Valery
> >> -------------------------
> >>> FROM: "Marco Mirazita" <Marco.Mirazita at lnf.infn.it>
> >>> TO: dsg-rich at jlab.org, rich at jlab.org
> >>> SENT: Tuesday, October 16, 2018 4:03:24 AM
> >>> SUBJECT: [Dsg-rich] FPGA temp alarm
> >>> Hi all,
> >>> yesterday at about 4:20 pm jlab time I received several alarm
> >>> messages
> >>> saying that the FPGA temperatures reached values around 115 deg.
> >>> Since it looked like a readout error, I checked the strip chart,
> >>> where
> >>> there is actually a spike in the readout at that time, but to
> > values
> >>> of
> >>> the order of -10^5. See the attached plot.
> >>> So, it is clear that it was a readout error, but it is strange that
> >>> the
> >>> alarm message and epics report different values.
> >>> Also, according to the logbook, the electronics was not shut down
> > by
> >>> the
> >>> interlocks, as should have happened if the temperature really went
> >>> above
> >>> 100 deg.
> >>> So, perhaps the values reported in the alarm messages are not the
> >>> correct ones?
> >>> Marco
> >>> _______________________________________________
> >>> Dsg-rich mailing list
> >>> Dsg-rich at jlab.org
> >>> https://mailman.jlab.org/mailman/listinfo/dsg-rich
> >> _______________________________________________
> >> Dsg-rich mailing list
> >> Dsg-rich at jlab.org
> >> https://mailman.jlab.org/mailman/listinfo/dsg-rich
> > _______________________________________________
> > Rich mailing list
> > Rich at jlab.org
> > https://mailman.jlab.org/mailman/listinfo/rich
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_rich/attachments/20181016/84b105d5/attachment-0002.html>
More information about the Dsg-rich
mailing list