denney,
with regards to your e-mail, much as i appreciate what you are saying, i'd like to remind you of an e-mail marc sent.
From: "Marc Mcmullen" <mcmullen at jlab.
To: "dsg-hallbgas" <dsg-hallbgas at jlab.org>, "dsg-rich" <dsg-rich at jlab.org>, "Denny Insley" <dinsley at jlab.org>
Cc: "Patrizia Rossi" <rossi at jlab.org>
Sent: Monday, May 7, 2018 3:09:14 PM
Subject: Re: [Dsg-hallbgas] [Dsg-internal] Pressure Systems work for Hall B N2 distribution system
Hi Denny and Bob,
Just a recap of this morning's conversation, DSG will need at least two weeks notice of any work which would cause an interruption of N2 to Hall B. This will give enough time to order high pressure LN2 dewars to be delivered to the hall. This will ensure that the Aerogel has a continuous source of N2 at all times to prevent the absorption of moisture.
Thanks,
Marc
also, this is not the first time such an incident has occured.
these are my recommendations:
1. anytime there is work scheduled that could affect the RICH, e-mail notice of work to dsg-rich at jlab.org.
2. with regards to developing a fail safe system for the N2, we discussed this in the DSG meeting, and george is developing such a system. the implementation of this system is up to Hall B and Hall B Engineering.
i recall on one occasion i told dr. rossi with regards to a certain issue:"insanity is repeating the same mistakes and expecting different results .", which is how i felt and feel.
her advise to me was "repitita iuvant-repetition does good"
so in this spirit, i'm repeating.
i hope that hallb/hallb eng will take the right approach.
amrit
From: "Denny Insley" <dinsley at jlab.org>
To: "Patrizia Rossi" <rossi at jlab.org>
Cc: "Amrit Yegneswaran" <yeg at jlab.org>, "Tyler Lemon" <tlemon at jlab.org>, "dsg-rich" <dsg-rich at jlab.org>, "Bob Miller" <rmiller at jlab.org>, "Marco Mirazita" <Marco.Mirazita at lnf.infn.it>
Sent: Wednesday, September 19, 2018 8:25:17 AM
Subject: Re: [Dsg-rich] URGENT: RICH alarm
Good morning All,
I am in complete agreement that this is excellent information but it brings to light the potential for major unexpected problems.
Historically the detectors, targets, magnets, vacuum systems and other components in Hall B have all been designed to fail "safe" in the
event of any system failure. Through interlocked valves, reliefs, circuit interrupts or other fast shut down devices, they are all designed
so that any foreseeable problem will be mitigated without the possibility of damaging the equipment. I am not in anyway trying to insinuate
that the RICH detector is an exception to that design but that it is a much more sensitive detector when it comes to operation. This is
evident from yesterday's interruption in nitrogen flow. Even though it was brief and controlled it obviously had a definitive impact on the detector.
Hall B's nitrogen supply is used in countless operations throughout the Hall and is frequently tapped into and used at a moments notice.
We have designed and are implementing a nitrogen supply panel in hopes of alleviating the "tapping into" of the nitrogen supply. The sole
source of nitrogen to the hall is the tank outside of the gas shed and every time that it is "tapped into" the overall integrity of the system
is weakened which in turn increases the potential for system failure. The majority of nitrogen use in the past has been purge and backfill
operations and therefore not a critical concern. The RICH detector is the first and only detector in the hall to be critically dependent on
nitrogen flow and requires all us to change our way of thinking when it comes to the importance of the supply. I would like to suggest that the
very first thing we do is add administrative tags to the supply valves going to the hall that contain contact and notification information,
instructing personnel to call prior to operating the valves. I also think that it is imperative to add an interlocked back up supply that will
take over in the case of a failure that can not be repaired in a very short period of time. In closing I am certain that with continued communication
and a solid team effort we can make this a very worry free system.
Thank you
Denny Insley
----- Original Message -----
From: "Patrizia Rossi" <rossi at jlab.org>
To: "Amrit Yegneswaran" <yeg at jlab.org>
Cc: "Tyler Lemon" <tlemon at jlab.org>, "dsg-rich" <dsg-rich at jlab.org>, "Bob Miller" <rmiller at jlab.org>, "Marco Mirazita" <Marco.Mirazita at lnf.infn.it>, "Denny Insley" <dinsley at jlab.org>
Sent: Tuesday, September 18, 2018 2:26:18 PM
Subject: Re: [Dsg-rich] URGENT: RICH alarm
yes, I agree with Amrit.
Thank you Tyler for the excellent analysis!
---------------------------------------------------
# Patrizia Rossi
# Jefferson Lab
# 12000 Jefferson Avenue
# Suite 5
# Newport News VA 23606 (USA)
# e-mail: rossi at jlab.org
# phone: +1 (757) 269 7740
# fax: + 1 (757) 269 5800
# cell: +1 (757) 528 7606
> On 18/set/2018, at 14:22, Amrit Yegneswaran <yeg at jlab.org> wrote:
>
> thanks for the excellent analysis tyler.
> this helps us set a baseline and endorses the fact that prior to scheduled work, we have to take appropriate action.
> for instance, the rich detector must have emergency nitrogen in the hall to switch to prior to scheduled work.
>
> detectors are like trust,
> takes years to build , seconds to destroy, and forever to repair.
>
>
> From: "Tyler Lemon" <tlemon at jlab.org>
> To: "Mirazita Marco" <Marco.Mirazita at lnf.infn.it>, "dsg-rich" <dsg-rich at jlab.org>
> Cc: "Robert Miller" <rmiller at jlab.org>, "Denny Insley" <dinsley at jlab.org>
> Sent: Tuesday, September 18, 2018 2:09:09 PM
> Subject: Re: [Dsg-rich] URGENT: RICH alarm
>
> Hello,
>
> In addition to the relative humidity sensors, we have the moisture transducer measuring the water concentration at the N2 Exhaust line.
>
> When N2 flow stopped there was 41.77 ppm of H2O.
> When N2 flow restarted, water concentration had increased to 46.72 ppm (4.95 ppm, or 11.86% overall, increase).
> One hour later, water concentration dropped to 44.17 ppm (2.55 ppm, or 5.46% overall, decrease).
>
> Attached is an updated table with the moisture transducer added in the last row in blue. Also attached is a plot during this time period for the moisture transducer.
>
> Best regards,
> Tyler
> From: "Tyler Lemon" <tlemon at jlab.org>
> To: "Mirazita Marco" <Marco.Mirazita at lnf.infn.it>, "dsg-rich" <dsg-rich at jlab.org>
> Cc: "Robert Miller" <rmiller at jlab.org>, "Denny Insley" <dinsley at jlab.org>
> Sent: Tuesday, September 18, 2018 1:44:49 PM
> Subject: Re: [Dsg-rich] URGENT: RICH alarm
>
> Hello,
>
> From the attached plot and table, it looks like during the time with no N2 flow, humidity increased by 0.28% RH (~ 30% overall increase from before flow loss).
>
> After flow had been restored for one hour, the humidity had decreased by 0.18% RH (~ 16% overall decrease from when flow was restored).
>
> Best regards,
> Tyler
>
> From: "Amrit Yegneswaran" <yeg at jlab.org>
> To: "Mirazita Marco" <Marco.Mirazita at lnf.infn.it>
> Cc: "dsg-rich" <dsg-rich at jlab.org>, "Robert Miller" <rmiller at jlab.org>, "Denny Insley" <dinsley at jlab.org>
> Sent: Tuesday, September 18, 2018 11:02:43 AM
> Subject: Re: [Dsg-rich] URGENT: RICH alarm
>
> denny, bob
> in future, for scheduled work please let the RICH group know before hand.
> amrit
>
> From: "Marco Mirazita" <Marco.Mirazita at lnf.infn.it>
> To: "Robert Miller" <rmiller at jlab.org>
> Cc: "dsg-rich" <dsg-rich at jlab.org>, "Denny Insley" <dinsley at jlab.org>
> Sent: Tuesday, September 18, 2018 10:56:28 AM
> Subject: Re: [Dsg-rich] URGENT: RICH alarm
>
> Hi all,
> it is back now.
> Marco
>
>
> Il 2018-09-18 16:54 Robert Miller ha scritto:
> > Hi Tyler,
> >
> > The N2 was stopped to add a valve; it will be back within the next 10
> > minutes.
> >
> > Bob
> >
> > Sent from my iPhone
> >
> > On Sep 18, 2018, at 10:51 AM, Tyler Lemon <tlemon at jlab.org> wrote:
> >
> >> Hello Bob and Denny,
> >>
> >> Is there any work being done on the forward carriage nitrogen supply
> >> line?
> >>
> >> RICH lost N2 flow and the aerogel needs to stay constantly purged by
> >> N2 to stay dry.
> >>
> >> Tyler
> >>
> >> -------------------------
> >>
> >> FROM: "Mirazita Marco" <Marco.Mirazita at lnf.infn.it>
> >> TO: rich at jlab.org, "dsg-rich" <dsg-rich at jlab.org>
> >> SENT: Tuesday, September 18, 2018 10:39:21 AM
> >> SUBJECT: [Dsg-rich] URGENT: RICH alarm
> >>
> >> Hi all,
> >> it seems that few minutes ago the N2 flow dropped to 0.
> >> Can someone please check what is going on?
> >> Thank you
> >> Marco
> >>
> >> _______________________________________________
> >> Dsg-rich mailing list
> >> Dsg-rich at jlab.org
> >> https://mailman.jlab.org/mailman/listinfo/dsg-rich [1]
> >
> >> <n2flow.png>
> >
> >
> > Links:
> > ------
> > [1] https://mailman.jlab.org/mailman/listinfo/dsg-rich
> >
> > _______________________________________________
> > Dsg-rich mailing list
> > Dsg-rich at jlab.org
> > https://mailman.jlab.org/mailman/listinfo/dsg-rich
> _______________________________________________
> Dsg-rich mailing list
> Dsg-rich at jlab.org
> https://mailman.jlab.org/mailman/listinfo/dsg-rich
>
> _______________________________________________
> Dsg-rich mailing list
> Dsg-rich at jlab.org
> https://mailman.jlab.org/mailman/listinfo/dsg-rich
>
>
> _______________________________________________
> Dsg-rich mailing list
> Dsg-rich at jlab.org
> https://mailman.jlab.org/mailman/listinfo/dsg-rich
>
> _______________________________________________
> Dsg-rich mailing list
> Dsg-rich at jlab.org
> https://mailman.jlab.org/mailman/listinfo/dsg-rich
> _______________________________________________
> Dsg-rich mailing list
> Dsg-rich at jlab.org
> https://mailman.jlab.org/mailman/listinfo/dsg-rich
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_rich/attachments/20180920/27330d35/attachment-0002.html>
More information about the Dsg-rich