[Dsg-rich] [New Logentry] RICH compressor pressure decrease investigation

Amrit Yegneswaran yeg at jlab.org
Tue Apr 9 18:20:13 EDT 2019


excellent tyler and co.


________________________________
From: Dsg-rich <dsg-rich-bounces at jlab.org> on behalf of tlemon at jlab.org <tlemon at jlab.org>
Sent: Tuesday, April 9, 2019 5:30 PM
To: dsg-rich at jlab.org
Subject: [Dsg-rich] [New Logentry] RICH compressor pressure decrease investigation

RICH compressor pressure decrease investigation<https://logbooks.jlab.org/entry/3677358>

  *   Edit<https://logbooks.jlab.org/entry/3677358/edit?destination=email/send>
  *   Delete<https://logbooks.jlab.org/entry/3677358/delete?destination=email/send>

Lognumber 3677358<https://logbooks.jlab.org/entry/3677358>. Submitted by tlemon<https://logbooks.jlab.org/user/tlemon> on Tue, 04/09/2019 - 17:27<https://logbooks.jlab.org/entries?start_date=1554841638&end_date=1554848838&book=RICH>.

Logbooks:       RICH<https://logbooks.jlab.org/book/rich>
Entry Makers:   beng, mcmullen
Attached Files:
[PDF icon]RICH-air-compressor-investigation-photos_2019-04-09.pdf<https://logbooks.jlab.org/files/2019/04/3677358/RICH-air-compressor-investigation-photos_2019-04-09.pdf>


Valery called notifying of low air pressure alarms for RICH's air cooling system. RICH air cooling buffer tank pressure had been steadily decreasing and was alarming at ~30 psi.

At Valery's direction, we made an access to the hall to investigate the cause.

In the hall, we noted that both the local pressure gauge on the cooling system buffer tank and the pressure readout on the compressor read ~80 psi (Fig 1 and Fig 2 in attachment). Because these two readouts match, it appears that the pressure transducer used for remote monitoring is not working correctly.

We left the compressor running and the system as we found it because:
1.) the cooling panel's regulators are set to lower than 80 psi (~35 psi for airflow 1 and ~30 psi for airflow 2; Fig 4 in attachment)
2.) airflows are close to their 500 slm and 400 slm set points (Fig 3 in attachment)
3.) the compressor's output pressure readback plot shows the pressure has been stable at ~80 for several days (Fig 2 in attachment).

For now, Valery will lower the air pressure alarm and trip level to avoid the system giving further alarms or tripping off the detector. Airflow interlocks will remain at the same level.

After the run, we will look into replacing the pressure transducer on the buffer tank and look further into why the compressor output pressure has dropped.

As a side note, we also noticed the gauge on the compressor's output filter was in the red (Fig 5 in attachment). This will also be investigated after the run to see if the filter needs to be drained.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_rich/attachments/20190409/38206b32/attachment-0002.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: application-pdf.png
Type: image/png
Size: 346 bytes
Desc: application-pdf.png
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_rich/attachments/20190409/38206b32/attachment-0002.png>


More information about the Dsg-rich mailing list